I have an Optiv Performance 1z443, a dual vision/touch system.
We were given a standard from Hexagon; it's a stand, with a glass calibration plate, a 20mm ring, and a gage ball artifact screwed into it.
Here's my process for calibrating (are you familiar with standard calibration procedures?)
First, I calibrate the camera probe offset on the ring, choosing YES the tool has moved.
Next, I'll calibrate a probe I don't use for anything else, a 4x10, on the ring, RESET TO THEO'S, choosing NO the tool has not moved.
Using that same probe, after calibrating on the ring, I will calibrate the sphere, DON'T RESET TO THEO'S, choose YES the tool has moved.
After that, calibrate every tip on the sphere, RESET TO THEO'S, choose NO the tool hasn't moved