Greetings,


I have been using the VB.Script ".GetFieldValue" since version 2014.1 to transfer measured values ​​to an external application.
This function ".GetFieldValue" ​​was nice to transfer data without an additional assign command

Unfortunately this no longer works with the new GeoTol in 2020R2.
There are simply no measured values ​​stored in GeoTol.Commands or I cannot find them.

Can someone tell me how else to access this data, if possible with a single function?










  • Greetings,

    i found a compromise. It is possible to read out the measured values ​​from the EditWindow.
    Because since 2020R2 the measured values ​​are also displayed there.

    don't get me wrong but this is a wheelchair with different sized tires.
    I would still be happy about a correct solution.


    But with this script, the measured values ​​of GeoTol can be extracted at runtime without additional commands
    This script probably only works with the English language setting because it parses the command text directly.
    (testet with English and German)


    Sub Main()
    On Error Resume Next
    
    ' Dim something
    Dim sOutput As String
    Dim vOutput As Object
    Set vOutput = CreateObject("PCDLRN.DimData")
    
    Dim CmdText, StrValue, SubCmdText, NumTest As String
    Dim nPos1, nPos2 As Integer
    Dim DblTest1, DblTest2 As Double
    NumTest = "0123456789"
    
    Dim App, Part, Cmds, Cmd As Object
    Set App = CreateObject("PCDLRN.Application")
    If (Not App.WaitUntilReady(300)) Or (App Is Nothing) Then
    MsgBox "Machine did not initialize, Exiting"
    Exit Sub
    End If
    Set Part = App.ActivePartProgram
    Set Cmds = Part.Commands
    Set Cmd = Nothing
    
    
    
    ' Loop all Commands
    For Each Cmd In Cmds
    ' Get Command Text from EditWindow
    CmdText = Cmd.Application.ActivePartProgram.EditWindow.GetCommandText(Cmd)
    
    
    ' only Geometric_Tolerace
    If (Cmd.IsDimension) And (InStr(1, CmdText, "=GEOMETRI") <> 0) Then
    
    ' Parse CmdText To OTOL
    'SEGMENT
    nPos1 = InStr(1, CmdText, "SEGMENT")
    vOutput.Plus = 0
    If nPos1 > 0 Then
    nPos2 = InStr(nPos1 + 1, CmdText, ":")
    SubCmdText = Mid(CmdText, nPos1 + 1, nPos2 - nPos1 - 1)
    
    nPos1 = InStr(1, SubCmdText, ",")
    nPos2 = InStr(nPos1 + 1, SubCmdText, ",")
    While nPos1 <> 0
    StrValue = Mid(SubCmdText, nPos1 + 1, nPos2 - nPos1 - 1)
    If InStr(1, NumTest, Left(StrValue, 1)) <> 0 Then
    vOutput.Plus = CDbl(StrValue)
    End If
    nPos1 = nPos2
    nPos2 = InStr(nPos1 + 1, SubCmdText, ",")
    Wend
    End If
    
    ' Parse CmdText To MEASURED
    'MULT=
    ':
    ': ,
    nPos1 = InStr(1, CmdText, "MULT=")
    vOutput.Meas = 0
    If nPos1 > 0 Then
    nPos2 = InStr(nPos1 + 1, CmdText, ":")
    If nPos2 > 0 Then
    nPos1 = InStr(nPos2 + 1, CmdText, ":")
    nPos2 = InStr(nPos1 + 1, CmdText, ",")
    
    While nPos1 <> 0
    StrValue = Mid(CmdText, nPos1 + 1, nPos2 - nPos1 - 1)
    If InStr(1, NumTest, Left(StrValue, 1)) <> 0 Then
    If vOutput.Meas = 0 Then
    ' Single Line MEASURED
    vOutput.Meas = CDbl(StrValue)
    vOutput.Max = vOutput.Meas
    vOutput.Min = vOutput.Meas
    Else
    ' special Case multi Line MEASURED
    If CDbl(StrValue) > vOutput.Max Then vOutput.Max = CDbl(StrValue)
    If CDbl(StrValue) < vOutput.Min Then vOutput.Min = CDbl(StrValue)
    End If
    End If
    nPos1 = InStr(nPos2 + 1, CmdText, ":")
    nPos2 = InStr(nPos1 + 1, CmdText, ",")
    Wend
    End If
    End If
    
    ' the rest
    vOutput.Bonus = 0
    vOutput.nom = 0
    vOutput.Minus = 0
    vOutput.Dev = vOutput.Meas
    vOutput.DevAngle = 0
    DblTest1 = vOutput.Plus
    DblTest2 = vOutput.Max
    If (DblTest1 - DblTest2) < 0 Then
    vOutput.Out = Abs(DblTest1 - DblTest2)
    Else
    vOutput.Out = 0
    End If
    
    
    ' Do something With vOutput
    sOutput = Cmd.ID & Chr(10) & Chr(13)
    sOutput = sOutput & "nominal:" & CStr(vOutput.nom) & Chr(10) & Chr(13)
    sOutput = sOutput & "measured: " & CStr(vOutput.Meas) & Chr(10) & Chr(13)
    sOutput = sOutput & "tol plus: " & CStr(vOutput.Plus) & Chr(10) & Chr(13)
    sOutput = sOutput & "tol minus: " & CStr(vOutput.Minus) & Chr(10) & Chr(13)
    sOutput = sOutput & "out of Tol: " & CStr(vOutput.Out) & Chr(10) & Chr(13)
    sOutput = sOutput & "max: " & CStr(vOutput.Max) & Chr(10) & Chr(13)
    sOutput = sOutput & "min: " & CStr(vOutput.Min)
    
    MsgBox sOutput
    
    End If
    Next Cmd
    
    
    Set Cmds = Nothing
    Set Cmd = Nothing
    Set Part = Nothing
    Set App = Nothing
    
    Set vOutput = Nothing
    End Sub
    

    Attached Files
  • Good Day,

    Finally, i figured out how it supposed to be done:

    With this you can call up the measured values ​​from the geoTol commands (2020R2) in a script and use them for something.
    (for example database storage)


    Sub test()
    
      Dim App, Part, Cmds, DmisCommand As Object
      Dim OutputText, sPuffer As String
      Dim RetVal
      Dim LoopIndex As Integer
    
      Set App = CreateObject("PCDLRN.Application")
      Set Part = App.PartPrograms.Item(1)
      Set Cmds = Part.Commands
    
      For Each DmisCommand In Cmds
    
        If (DmisCommand.Type = ISO_TOLERANCE_COMMAND) Or (DmisCommand.Type = ASME_TOLERANCE_COMMAND) Then
          OutputText = "STANDARD: " & DmisCommand.GetText(STANDARD, 0) & Chr(13)
          OutputText = OutputText & "UNIT_TYPE: " & DmisCommand.GetText(UNIT_TYPE, 0) & Chr(13)
          OutputText = OutputText & "SEGMENT_TYPE_TOGGLE: " & DmisCommand.GetText(SEGMENT_TYPE_TOGGLE, 1) & Chr(13)
          OutputText = OutputText & "OUTPUT_TYPE: " & DmisCommand.GetText(OUTPUT_TYPE, 0) & Chr(13)
          OutputText = OutputText & "ARROW_DENSITY: " & DmisCommand.GetText(ARROW_DENSITY, 0) & Chr(13)
          OutputText = OutputText & "Upper Toleranz: " & DmisCommand.GetText(FORM_TOLERANCE, 1) & Chr(13)
          OutputText = OutputText & "lower Toleranz: " & "0" & Chr(13)
    
    
          LoopIndex = 1
          sPuffer = DmisCommand.GetText(REF_ID, LoopIndex)
          While sPuffer <> ""
            OutputText = OutputText & " ->" & sPuffer & " = " & DmisCommand.GetTextEx(DIM_DEVIATION, LoopIndex, "SEG=1") & Chr(13)
    
            LoopIndex = LoopIndex + 1
            sPuffer = DmisCommand.GetText(REF_ID, LoopIndex)
          Wend
    
          MsgBox OutputText
        End If
    
      Next DmisCommand
    End Sub
    
  • God damn it, you had me worried there... Smiley
  • From the top example how do you get the diameter to get output as it only outputs the data regarding the position of a hole for example...
  • good day,

    i assume you looking for this command
    this only works if there is a reference with a diameter

    getting the nominal of the reference ID
    DmisCommand.GetTextEx(SIZE_NOMINAL, 0, "SIZE")



    getting the mesured of the reference ID
    DmisCommand.GetText(UPPER_SIZE, LoopIndex)

    DmisCommand.GetText(LOWER_SIZE, LoopIndex)


    Sub test()
      Dim App, Part, Cmds, DmisCommand As Object
      Dim OutputText, sPuffer As String
      Dim RetVal
      Dim LoopIndex As Integer
    
      Set App = CreateObject("PCDLRN.Application")
      Set Part = App.PartPrograms.Item(1)
      Set Cmds = Part.Commands
    
    
      For Each DmisCommand In Cmds
    
      If (DmisCommand.Type = ISO_TOLERANCE_COMMAND) Or (DmisCommand.Type = ASME_TOLERANCE_COMMAND) Then
        OutputText = "STANDARD: " & DmisCommand.GetText(STANDARD, 0) & Chr(13)
        OutputText = OutputText & "UNIT_TYPE: " & DmisCommand.GetText(UNIT_TYPE, 0) & Chr(13)
        OutputText = OutputText & "SEGMENT_TYPE_TOGGLE: " & DmisCommand.GetText(SEGMENT_TYPE_TOGGLE, 1) & Chr(13)
        OutputText = OutputText & "OUTPUT_TYPE: " & DmisCommand.GetText(OUTPUT_TYPE, 0) & Chr(13)
        OutputText = OutputText & "Upper Toleranz: " & DmisCommand.GetText(FORM_TOLERANCE, 1) & Chr(13)
        OutputText = OutputText & "lower Toleranz: " & "0" & Chr(13)
        OutputText = OutputText & "SIZE_NOMINAL: " & DmisCommand.GetTextEx(SIZE_NOMINAL, 0, "SIZE") & Chr(13)
    
        LoopIndex = 1
        sPuffer = DmisCommand.GetText(REF_ID, LoopIndex)
        While sPuffer <> ""
          OutputText = OutputText & " ->" & sPuffer & " DIM_DEVIATION = " & DmisCommand.GetTextEx(DIM_DEVIATION, LoopIndex, "SEG=1") & Chr(13)
          OutputText = OutputText & " ->" & sPuffer & " UPPER_SIZE = " & DmisCommand.GetText(UPPER_SIZE, LoopIndex) & Chr(13)
          OutputText = OutputText & " ->" & sPuffer & " LOWER_SIZE = " & DmisCommand.GetText(LOWER_SIZE, LoopIndex) & Chr(13)
          OutputText = OutputText & " " & Chr(13)
    
          LoopIndex = LoopIndex + 1
          sPuffer = DmisCommand.GetText(REF_ID, LoopIndex)
        Wend
    
    
        MsgBox OutputText
      End If
    
      Next DmisCommand
    End Sub
    
  • Thanks I will make sure Simon see's this when he gets returns from his holiday
  • That's great that works a treat...I can incorporate that into my existing code.....Thank you so much for this....
  • I am not experienced with coding except for the online coding class Hexagon has. Where would this additional scrip go or what would it replace in the script I have? This script is a script they were using here before I started.

    As it is right now, in the excel report it recognized the dimension number but not the nominal/tolerance/result. This is causing us to have to go back and enter these values in by hand which is very time consuming. I also have the Excel template if its needed to help figure this out.

    I know I am asking someone to figure this out for me and I hate doing that. However, at this point in time, I need this fixed and don't have the time to figure it out myself.

    Sub Main (strVariable As String, reasonVar As String)
    
    
    'xl Declarations
    Dim xlApp As Object
    Dim xlWorkbooks As Object
    Dim xlWorkbook As Object
    Dim xlSheet As Object
    Dim count As Integer
    
    
    'pcdlrn declarations And Open ppg
    Dim App As Object
    Set App = CreateObject("PCDLRN.Application")
    Dim Part As Object
    Set Part = App.ActivePartProgram
    Dim Cmds As Object
    Set Cmds = Part.Commands
    Dim Cmd As Object
    Dim DCmd As Object
    Dim DcmdID As Object
    Dim DimID As String
    Dim fs As Object
    Dim ReportDim As String
    Dim CheckDim As String
    
    
    'Check To see If results file exists
    FilePath = "" '.xlsm And .bas files location
    DataPath = "" 'report save location
    Set fs = CreateObject("Scripting.FileSystemObject")
    ResFileExists = fs.fileexists(DataPath & Part.partname & " " & strVariable & " " & reasonVar & ".xlsm") 'check program folder For .xlsm file
    
    
    'Open Excel And Base form
    Set xlApp = CreateObject("Excel.Application")
    Set xlWorkbooks = xlapp.Workbooks
    If ResFileExists = False Then
    TempFilename = FilePath & "Loop Template Column.xlsm"
    Else
    TempFilename = DataPath & Part.partname & " " & strVariable & " " & reasonVar & ".xlsm"
    End If
    Set xlWorkbook = xlWorkbooks.Open(TempFilename)
    Set xlSheet = xlWorkbook.Worksheets("Sheet1")
    
    
    If ResFileExists = False Then
    RCount=6
    CCount=3
    xlSheet.Range("B1").Value = Part.PartName
    xlSheet.Range("E4").Value = Date() & " " & Time()
    xlSheet.Range("D1").Value = strVariable
    xlSheet.Range("C2").Value = reasonVar
    
    
    
    For Each Cmd In Cmds
    'Eliminate DATDEF's
    If Cmd.Type <> 1299 Then
    'Do Dimensions
    If Cmd.IsDimension Then
    If Cmd.Type = DIMENSION_START_LOCATION Or Cmd.Type = DIMENSION_TRUE_START_POSITION Then
    Set DcmdID = Cmd.DimensionCommand
    DimID = DcmdID.ID
    ReportDim = Cmd.GetText ( OUTPUT_TYPE, 0)
    End If
    If Cmd.Type <> DIMENSION_START_LOCATION And Cmd.Type <> DIMENSION_END_LOCATION And _
    Cmd.Type <> DIMENSION_TRUE_START_POSITION And Cmd.Type <> DIMENSION_TRUE_END_POSITION Then
    Set DCmd = Cmd.DimensionCommand
    CheckDim = Cmd.GetText ( OUTPUT_TYPE, 0)
    If CheckDim <> "" Then
    ReportDim = CheckDim
    End If
    If ReportDim = "BOTH" Or ReportDim = "STATS" Then
    If DCmd.ID = "" Then
    xlSheet.Cells(RCount,4).Value = DimID & " . "& DCmd.AxisLetter
    Else
    xlSheet.Cells(RCount,4).Value = DCmd.ID & " . " & "M"
    End If
    xlSheet.Cells(RCount,1).Value = DCmd.Nominal
    xlSheet.Cells(RCount,2).Value = DCmd.Plus
    xlSheet.Cells(RCount,3).Value = DCmd.Minus
    'Measured Or Deviation With check For True Position
    If DCmd.AxisLetter <> "TP" Then
    xlSheet.Cells(RCount,5).Value = DCmd.Measured
    Else
    xlSheet.Cells(RCount,5).Value = DCmd.Deviation
    End If
    'Add Min/Max For Profile dimensions
    If Cmd.Type = 1118 Or Cmd.Type = 1105 Then
    RCount=RCount+1
    xlSheet.Cells(RCount,4).Value = DCmd.ID & "." & "Max"
    xlSheet.Cells(RCount,1).Value = DCmd.Nominal
    xlSheet.Cells(RCount,2).Value = DCmd.Plus
    xlSheet.Cells(RCount,3).Value = DCmd.Minus
    xlSheet.Cells(RCount,5).Value = DCmd.Max
    RCount=RCount+1
    xlSheet.Cells(RCount,4).Value = DCmd.ID & "." & "Min"
    xlSheet.Cells(RCount,1).Value = DCmd.Nominal
    xlSheet.Cells(RCount,2).Value = DCmd.Plus
    xlSheet.Cells(RCount,3).Value = DCmd.Minus
    xlSheet.Cells(RCount,5).Value = DCmd.Min
    End If
    RCount=RCount+1
    End If
    End If
    End If
    'Do GDT
    If Cmd.Type = 184 Then
    ReportDim = Cmd.GetText ( OUTPUT_TYPE, 0)
    If ReportDim = "BOTH" Or ReportDim = "STATS" Then
    xlSheet.Cells(RCount,4).Value = Cmd.GetText (ID, 0) & "." & "FCF"
    xlSheet.Cells(RCount,1).Value = "0"
    xlSheet.Cells(RCount,2).Value = Cmd.GetText (LINE2_PLUSTOL, 1)
    xlSheet.Cells(RCount,3).Value = "0"
    xlSheet.Cells(RCount,5).Value = Cmd.GetText (LINE2_DEV, 1)
    RCount=RCount+1
    End If
    End If
    End If
    Next Cmd
    
    
    Else
    
    'Find first Open column.
    CCount=5
    Found=0
    Do Until Found = 1
    CCount = CCount + 1
    If xlSheet.Cells(4,CCount).Value = "" Then
    Found=1
    End If
    Loop
    
    xlSheet.Cells(4,CCount).Value = Date() & " " & Time()
    xlSheet.Cells(5,CCount).Value = " Part " & CCount - 4
    
    'Fill In measured data
    RCount = 6
    For Each Cmd In Cmds
    'Eliminate DATDEF's
    If Cmd.Type <> 1299 Then
    'Do Dimensions
    If Cmd.IsDimension Then
    If Cmd.Type = DIMENSION_START_LOCATION Or Cmd.Type = DIMENSION_TRUE_START_POSITION Then
    Set DcmdID = Cmd.DimensionCommand
    DimID = DcmdID.ID
    ReportDim = Cmd.GetText ( OUTPUT_TYPE, 0)
    End If
    If Cmd.Type <> DIMENSION_START_LOCATION And Cmd.Type <> DIMENSION_END_LOCATION And _
    Cmd.Type <> DIMENSION_TRUE_START_POSITION And Cmd.Type <> DIMENSION_TRUE_END_POSITION Then
    Set DCmd = Cmd.DimensionCommand
    CheckDim = Cmd.GetText ( OUTPUT_TYPE, 0)
    If CheckDim <> "" Then
    ReportDim = CheckDim
    End If
    If ReportDim = "BOTH" Or ReportDim = "STATS" Then
    'Measured Or Deviation With check For True Position
    If DCmd.AxisLetter <> "TP" Then
    xlSheet.Cells(RCount,CCount).Value = DCmd.Measured
    Else
    xlSheet.Cells(RCount,CCount).Value = DCmd.Deviation
    End If
    'Add Min/Max For Profile dimensions
    If Cmd.Type = 1118 Or Cmd.Type = 1105 Then
    RCount=RCount+1
    xlSheet.Cells(RCount,CCount).Value = DCmd.Max
    RCount=RCount+1
    xlSheet.Cells(RCount,CCount).Value = DCmd.Min
    End If
    Rcount=Rcount+1
    End If
    End If
    End If
    'Do GDT
    If Cmd.Type = 184 Then
    ReportDim = Cmd.GetText ( OUTPUT_TYPE, 0)
    If ReportDim = "BOTH" Or ReportDim = "STATS" Then
    xlSheet.Cells(RCount,CCount).Value = Cmd.GetText (ID, 0) & "." & "FCF"
    xlSheet.Cells(RCount,CCount).Value = "0"
    xlSheet.Cells(RCount,CCount).Value = Cmd.GetText (LINE2_PLUSTOL, 1)
    xlSheet.Cells(RCount,CCount).Value = "0"
    xlSheet.Cells(RCount,CCount).Value = Cmd.GetText (LINE2_DEV, 1)
    RCount=RCount+1
    End If
    End If
    End If
    Next Cmd
    End If
    
    
    'Save And Cleanup
    Set xlSheet = Nothing
    SaveName = DataPath & Part.partname & " " & strVariable & " " & reasonVar & ".xlsm"
    If ResFileExists = False Then
    xlWorkBook.SaveAs SaveName
    Else
    xlWorkBook.Save
    End If
    xlWorkbook.Close
    Set xlWorkbook = Nothing
    xlWorkbooks.Close
    Set xlWorkbooks = Nothing
    xlApp.Quit
    Set xlApp = Nothing
    
    LabelEnd:
    
    End Sub
  • this should be no problem,
    give me some time
  • hi



    i have modified your basic Script " ExtractGEOv2.txt "

    i dont have a Machine free so i couldn't test it