For anyone out there with an LSP X1h and are using 2015.0 or 2015.1:
Experienced any problems with calibrating your probes?

Example:

For the first two probes below, I have answered YES to the question if the tool has moved (only way in order to get values that matches the THEOs):
T1A-5B-15      TEO X  -1.5046 Y  -5.6152 Z 284.6301 D 5.0000
T1A-5B-15      SNABB X  -1.4635 Y  -5.5247 Z 284.6052 D 5.0000 PrbRdv 0.0035
T1A-5B-15      MÄTT X  -1.4635 Y  -5.5248 Z 284.6044 D 5.0000 PrbRdv-0.0012 StdDev 0.0004

T1A0B0         TEO X   0.0000 Y  12.0000 Z 284.3600 D 5.0000
T1A0B0         SNABB X   0.0016 Y  11.9994 Z 284.3585 D 5.0000 PrbRdv 0.0031
T1A0B0         MÄTT X   0.0014 Y  11.9994 Z 284.3579 D 5.0000 PrbRdv-0.0017 StdDev 0.0004

T1A-90B-90     TEO X-203.8600 Y  -0.0000 Z  92.5000 D 5.0000
T1A-90B-90     SNABB X-203.0024 Y   0.0033 Z  93.3571 D 5.0000 PrbRdv 0.0027
T1A-90B-90     MÄTT X-203.0017 Y   0.0032 Z  93.3571 D 5.0000 PrbRdv-0.0015 StdDev 0.0005

T1A-90B90      TEO X 203.8600 Y  -0.0000 Z  92.5000 D 5.0000
T1A-90B90      SNABB X 202.5958 Y   0.4775 Z  93.2956 D 5.0000 PrbRdv 0.0034
T1A-90B90      MÄTT X 202.5957 Y   0.4776 Z  93.2956 D 5.0000 PrbRdv-0.0009 StdDev 0.0004

T1A90B-180     TEO X   0.0000 Y-203.8600 Z  68.5000 D 5.0000
T1A90B-180     SNABB X  -0.3890 Y-202.5555 Z  69.5875 D 5.0000 PrbRdv 0.0026
T1A90B-180     MÄTT X  -0.3891 Y-202.5557 Z  69.5873 D 5.0000 PrbRdv-0.0015 StdDev 0.0004

T1A90B0        TEO X   0.0000 Y 203.8600 Z  68.5000 D 5.0000
T1A90B0        SNABB X  -0.0159 Y 203.0495 Z  69.6721 D 5.0000 PrbRdv 0.0033
T1A90B0        MÄTT X  -0.0159 Y 203.0492 Z  69.6725 D 5.0000 PrbRdv-0.0015 StdDev 0.0005


The machine can't complete a DCC calibration of the angles, only A0B0... It either touches the calibration sphere or errors out giving me the "Invalid reading of measured point".

Hexagon tech tested the calibration using 2013 MR1 which passed, ie. PC-DMIS could calibrate the angles as expected so it seems that the issue is related to 2015.0 AND 2015.1. I have tried with deleting the probe files, creating new ones, creating new tool, hit/retract values, analog probe values etc., to no avail. Needless to say, machine is not fit for use in it's current state - if used together with 2015.*

Why not downgrade? Because with our CMM programs collection, backsaving is out of the question - that's why.

Anyone else seen this?
  • Just some thoughts... :
    Angle offsets can be far from meas to theo because of a bad orientation of the tesastar/ ph10.
    About lower matrix, vpt.se said that he can calibrate A0B0, so he can do it (and I think he did !)
    On new controllers, is the lower matrix is in the controller (and stays on) or is it just send from the settings / register ?

    What about dcc+dcc calibration ? Does it change anything ?
  • Just some thoughts... :
    Angle offsets can be far from meas to theo because of a bad orientation of the tesastar/ ph10.


    That's a valuable thought. Definitely something to consider.
  • That's a valuable thought. Definitely something to consider.


    Thx !
    In this case, a Man+DCC calibration should solve the problem...
  • "Invalid reading of measured point" is a tricky one, a message that pops up for several intermittent probing issues.

    One common cause is a slightly incorrect probe build (real world probe sightly longer than virtual build) so the stylus makes contact before it's fully deaccelerated from MOVESPEED down to TOUCHSPEED.

    I skimmed the thread, did we rule out improper Opt Probe settings in PRG?
  • Is this your first FDC controller or do you have others working correctly in the shop? I ask because FDC handles settings completely differently than common controller. When you execute a program with an OP Probe command it passes these values to the controller and they remain active for the duration of that session but they are not saved. If you turn the controller off/on those settings are gone. So you must have the Opt probe command in every program to make sure you are always passing the correct settings to the controller. If you don't you could end up running with controller defaults which are not correct.

    Common controllers saved the settings so they would persist even through a power cycle. This made it unnecessary to have the opt probe in every program unless you were specifically trying to change something.
  • What about dcc+dcc calibration ? Does it change anything ?


    Only that it completes the calibration, ie. it doesn't stop due to error.

    "Invalid reading of measured point" is a tricky one, a message that pops up for several intermittent probing issues.

    One common cause is a slightly incorrect probe build (real world probe sightly longer than virtual build) so the stylus makes contact before it's fully deaccelerated from MOVESPEED down to TOUCHSPEED.

    I skimmed the thread, did we rule out improper Opt Probe settings in PRG?


    Measured the actual probe/stylus and it matches the NAME of the probe, haven't checked actual probe definition in probe file. No additional probe settings/commands in the program.

    Is this your first FDC controller or do you have others working correctly in the shop? I ask because FDC handles settings completely differently than common controller. When you execute a program with an OP Probe command it passes these values to the controller and they remain active for the duration of that session but they are not saved. If you turn the controller off/on those settings are gone. So you must have the Opt probe command in every program to make sure you are always passing the correct settings to the controller. If you don't you could end up running with controller defaults which are not correct.

    Common controllers saved the settings so they would persist even through a power cycle. This made it unnecessary to have the opt probe in every program unless you were specifically trying to change something.


    This is the only FDC box we have. No additional probe commands in the program. Settings/values checked by Hexagon tech, I have even changed the values for the LSP X1h to the values recommended by Hexagon, to no avail. Tech brought a new LSP X1h, performed the same as the old one ("our" X1). Tesahead checked by tech.
  • Hi VPT !
    Could you try to calibrate a star probe, just to check if it's a rotation problem or a soft problem ?
    Have a good thursday ! (even if it's hard to look at the solution !!!!!)
  • An odd thing is that using a retract distance of 0.5 mm during calibration yields the "invalid reading of the measured point". Using a 2 mm retract distance and the machine might pass through calibration. I just tested a star probe calibration (2 mm retract) and the A0B0 passes just fine, but when moving to the angled positions (A90B-90), I get the "Probe deflection during CNC movement".
  • Vpt, a technician is planned to visit you on Monday. Also please check the things I outlined in mail.
  • In my experience many techs do not yet have a good understanding of the FDC controllers. Could very well be that the first technician missed something. Sounds like Anders has someone heading your way so hopefully theyll get you sorted out.