Our 1x50mm probe is showing a stddev of .0006 and can show up to 0.001 at times.
Thanks!
If you are using carbon fiber anywhere in your stylus build you might be seeing the effects of micro-fractures in the carbon fiber.
Usually if the SM module is going bad you see evidence of the sprung spring as previously mentioned and/or you fail the qualification routine.
It sounds like you have mulitple modules machines & stylii, so check them against each other. Run part #1 on CMM #1 with SM 1 module #1 get results that show profile out - conflicting with other methods showing profile to be in. Next: Run part #1 on CMM #1 with SM 1 module #2 get results that show profile in - matching other method results = SM 1 # 1 is bad.
Repeat: exchanging a single component at a time until you find the culprit. If you think a carbon fiber extension is suspect, do as above but change out the stylus rather than the SM 1.
Basic methodical, scientific, approach to trouble shooting. May it serve you well.
HTH
When you say your seeing errors on scans, are you using a cad model and did you check the surface vectors of the model.
I've had problems with this before, especially with V2010 mr3 which I see you are using.
So by saying you are a tech does this mean you don't have the ability to manipulate the programs being run? Are scans only OOT at certain parts of the scan? Perhaps at a curved surface or maybe a flat? If so and if you have the ability have you tried to take a single point at the affected area? Did it match the result of the scan? How tight of tolerances are we talking about and are you doing a form only or form and location? You also said these are known good parts - did these same parts originally check good via a scan or you hard checked them therefore know they are good?