The 2 machines are repeating but there is as John stated .0005" variance between the 2
try recalibrating with a lower touch speed, this looks like a calibration variable is cause
CMM A is off 0.010mm ?? start there---that being said ring is 117.5 do you have a cert? what does that state. That would be comparison value.
are these machines in an environmentally different area of shop? Had an issue with a correlation after a vent was altered for a machinist and changed to get cool air at him which also happened to be at the SF CMM.....caused a bit of a shift
I level these machine when moved---no one said do it just seemed to make sense....just another thought