The 2 machines are repeating but there is as John stated .0005" variance between the 2
try recalibrating with a lower touch speed, this looks like a calibration variable is cause
CMM A is off 0.010mm ?? start there---that being said ring is 117.5 do you have a cert? what does that state. That would be comparison value.
are these machines in an environmentally different area of shop? Had an issue with a correlation after a vent was altered for a machinist and changed to get cool air at him which also happened to be at the SF CMM.....caused a bit of a shift
I level these machine when moved---no one said do it just seemed to make sense....just another thought
Hurricane went more North of my location, I'm considering myself lucky. Never lost power or any utilities. Others not so much.
Ok, this is what I did.
RESULTS
- Created Test Program in MM
- Created Standard Force 4x20 (A0B0) (don't have a light force or 5x20)
- Created Cal Sphere in mm units (29.0504, all 6 digits entered in menu)
- Calibrated 25 hits at 4 levels at touchspeed of 5mm
- Used a 117.5mm master ring for measurement with touchspeed of 5mm throughout
- Used same probe, sphere, and ring for both machines.
- Calibration results on both machines were 0.002
- Created a model for autofeatures to be used during programming.
- Manual and DCC alignment, Z+ plane, XY Orgin cylinder
- Looped an auto circle with 23 pnts 5 times.
CMM_A
117.513
117.514
117.515
117.515
117.514
CMM_B
117.501
117.502
117.502
117.502
117.502