hexagon logo

"No probing within probelimit"

Hi, all. today i did run my calibration program and everyting seemed good, but after the first exchange of probes i reicive error message "TR_PRB01,no Probing within Probelimit !"
to me it`s odd, because i did run the calibration program for a while ago.

i tried to jump over the probe wich have this problem, and 3 of 6 probes are running throug without any problems.

so far i`ve been checking the probes and they are all in the same condition as it was the last time i performed a calibration.

if it helps, i run PC-DMIS 2016

in advance thank you all.
Parents
  • By "position" are you referring to the probe angle? I don't know of a way to automatically autocalibrate based on the program. I believe you will need to save that in the parameter set. When I said select the tips that would be Tip # and A&B angles, e.g. T1A0B0, T2A0B90, etc. Select the ones you want to calibrate and save them in a parameter set. Then you can call up that specific set with the AUTOCALIBRATE command. If you named the set CAL_1 then it would look like this.

    AUTOCALIBRATE/PROBE, PARAMETER_SET=CAL_1, QUALTOOL_MOVED=NO,
                              SHOW_SUMMARY=NO, OVERWRITE_RESULTSFILE=NO
Reply
  • By "position" are you referring to the probe angle? I don't know of a way to automatically autocalibrate based on the program. I believe you will need to save that in the parameter set. When I said select the tips that would be Tip # and A&B angles, e.g. T1A0B0, T2A0B90, etc. Select the ones you want to calibrate and save them in a parameter set. Then you can call up that specific set with the AUTOCALIBRATE command. If you named the set CAL_1 then it would look like this.

    AUTOCALIBRATE/PROBE, PARAMETER_SET=CAL_1, QUALTOOL_MOVED=NO,
                              SHOW_SUMMARY=NO, OVERWRITE_RESULTSFILE=NO
Children
No Data