can someone explain to me how exactly "Automatically Calibrate a Single Tip" works.
So far I have only managed that pcDMIS gives me an error message that the matrix cannot be calculated.
otherwise I've always used AutoCalibrate Probe with no problems.
I need a small program that calibrates 5 tips/positions each on three different probes without me being at the machine.
(the probes have many more positions)
I would be happy if someone could provide me with a working example routine or a piece of code.
Select the tip(s) you want
Click Measure
Define a Parameter set name and click Save
In the program use and Auto Calibrate command and select the parameter set name instead of All_TIPS_WITH_DEFAULTS.
You can have multiple AutoQual commands in you need to do a few tips per probe