CMM crashes on probe changer lids

Hello to you all,

I've recently installed the HR-MP 6 port probe changer on one of our CMMs. It was calibrated and successfully changed probes without issue. However, when I close the lids, instead of pushing the covers over when picking up a probe, it moves over slightly above the lid then crashes on top. I've tried playing with the (before and after pick-up distances) and (edit the probe data) to try and "force" the CMM to slide the lids over before picking up a probe. I would assume the probe changer should operate with the lids closed but for now I'm leaving them open. Is there anything I need to change to make this work with the lids closed?


Parents
  • Had a similar issue with whatever Hexagon calls their scanning probe setup. One day it just started crashing. Recalibrate, and it still crashes. Found the Z height for the pockets to be out the radius of the probe used to calibrate. Lowered the z position for all the pockets manually. Didn't have a problem for the remainder of my time there.
Reply
  • Had a similar issue with whatever Hexagon calls their scanning probe setup. One day it just started crashing. Recalibrate, and it still crashes. Found the Z height for the pockets to be out the radius of the probe used to calibrate. Lowered the z position for all the pockets manually. Didn't have a problem for the remainder of my time there.