I'm really close to ditching any attempt to use looping in my program to try and run multiple parts on my CMM table (I would like to run maybe 6 max). I'm trying to use 1 probe to inspect all the parts features (measuring on part, then going down in -Y 11.5" and doing the same measurement, go down in -Y again etc), then change probes, and measure all parts the same way (measuring all necessary features of one part, and then going down 11.5 inches in Y). The alignment keeps changing on me from the first part to the second...and when I post the data, 1 part will have the alignment from the second part. So, everything will dimensionally be okay, but because the origin is shifted down in Y by 11.5 inches, everything is screwed up on the total positioning (even though I know these parts will pass). How can I save one alignment/ do an external alignment, and then just shift it down 11.5 inches? I am sick of using looping, I feel like looping is really only for like bolt hole patterns, and waffle grid patterns in holes. Measuring whole parts in different locations is giving me a nightmare. It's something about the tool change and multiple parts. Because the second part doesn't get it's own alignment (like if your DCC alignment was A3, the second part should get A3-2 or something like that, but no. It just drives off the first parts alignment. I need multiple alignments just like the first part, just in different locations. I'll post the code just for giggles.
STARTUP =ALIGNMENT/START,RECALL:USE_PART_SETUP,LIST=YES
ALIGNMENT/END
MODE/MANUAL
FORMAT/TEXT,OPTIONS, ,HEADINGS,SYMBOLS, ;NOM,TOL,MEAS,DEV,OUTTOL, ,
LOADPROBE/4MM X 20MM
TIP/T1A0B0, SHANKIJK=0, 0, 1, ANGLE=0
PLN1 =FEAT/PLANE,CARTESIAN,TRIANGLE
THEO/<3.4829,-0.1456,0>,<0,0,1>
ACTL/<12.1376,32.7697,-26.1905>,<0.0005222,-0.0003643,0.9999998>
MEAS/PLANE,3
HIT/BASIC,NORMAL,<1.9889,-0.5175,0>,<0,0,1>,<10.5594,32.265,-26.1899>,USE THEO=YES
HIT/BASIC,NORMAL,<4.0258,0.8213,0>,<0,0,1>,<12.1295,33 .9516,-26.1901>,USE THEO=YES
HIT/BASIC,NORMAL,<4.434,-0.7405,0>,<0,0,1>,<13.7238,32.0923,-26.1916>,USE THEO=YES
ENDMEAS/
LIN1 =FEAT/LINE,CARTESIAN,UNBOUNDED
THEO/<-2.1728,-2.295,-0.6305>,<1,0,0>
ACTL/<6.7011,30.8644,-26.6467>,<1,-0.0003111,0>
MEAS/LINE,2,ZPLUS
HIT/BASIC,NORMAL,<-2.1728,-2.295,-0.629>,<0,-1,0>,<6.7011,30.8644,-26.6468>,USE THEO=YES
HIT/BASIC,NORMAL,<6.84,-2.295,-0.632>,<0,-1,0>,<15.6711,30.8616,-26.6467>,USE THEO=YES
ENDMEAS/
PNT1 =FEAT/POINT,CARTESIAN
THEO/<-2.275,0.0718,-0.3763>,<-1,0,0>
ACTL/<6.3863,31.8955,-26.567>,<-0.9999996,-0.0009195,0>
MEAS/POINT,1,WORKPLANE
HIT/BASIC,NORMAL,<-2.275,0.0718,-0.3763>,<-1,0,0>,<6.3863,31.8955,-26.567>,USE THEO=YES
ENDMEAS/
A1 =ALIGNMENT/START,RECALL:STARTUP,LIST=YES
ALIGNMENT/LEVEL,ZPLUS,PLN1
ALIGNMENT/TRANS,ZAXIS,PLN1
ALIGNMENT/ROTATE,XPLUS,TO,LIN1,ABOUT,ZPLUS
ALIGNMENT/TRANS,YAXIS,LIN1
ALIGNMENT/TRANS,XAXIS,PNT1
ALIGNMENT/END
MODE/DCC
FLY/ON
MOVESPEED/ 250
MOVE/POINT,NORMAL,<4.5328,1.9318,1.0961>
PLN2 =FEAT/PLANE,CARTESIAN,TRIANGLE
THEO/<5.7549,2.0759,0>,<0,0,1>
ACTL/<5.755,2.0756,-0.0033>,<0.0000146,-0.0000177,1>
MEAS/PLANE,3
MOVE/POINT,NORMAL,<7.0084,1.6156,0.8147>
(CODE CUT TO FIT, JUST MOVE POINTS)
HIT/BASIC,NORMAL,<4.2265,1.4394,0>,<0,0,1>,<4.2264,1.4 389,-0.0033>,USE THEO=YES
HIT/BASIC,NORMAL,<6.0296,3.0913,0>,<0,0,1>,<6.0298,3.0 911,-0.0033>,USE THEO=YES
HIT/BASIC,NORMAL,<7.0084,1.6972,0>,<0,0,1>,<7.0086,1.6 967,-0.0033>,USE THEO=YES
ENDMEAS/
MOVE/POINT,NORMAL,<6.9272,2.3614,1.4628>
MOVE/POINT,NORMAL,<6.8272,-0.616,0.8246>
MOVE/POINT,NORMAL,<0.1955,-0.616,1.1066>
MOVE/POINT,NORMAL,<0.1955,-0.3262,-0.6042>
LIN2 =FEAT/LINE,CARTESIAN,UNBOUNDED
THEO/<0.2769,0,-0.6506>,<1,0,0>
ACTL/<0.2769,0.0045,-0.6505>,<1,-0.0000029,0>
MEAS/LINE,2,ZPLUS
HIT/BASIC,NORMAL,<0.2769,0,-0.6573>,<0,-1,0>,<0.2769,0.0045,-0.6573>,USE THEO=YES
HIT/BASIC,NORMAL,<8.968,0,-0.6438>,<0,-1,0>,<8.9682,0.0045,-0.6437>,USE THEO=YES
ENDMEAS/
MOVE/POINT,NORMAL,<8.968,-0.4155,1.3796>
MOVE/POINT,NORMAL,<-0.591,-0.4155,1.5577>
MOVE/POINT,NORMAL,<-0.6887,-0.4155,-0.3356>
PNT2 =FEAT/POINT,CARTESIAN
THEO/<0,2.1944,-0.3618>,<-1,0,0>
ACTL/<0.0011,2.1942,-0.3618>,<-1,0,0>
MEAS/POINT,1,WORKPLANE
HIT/BASIC,NORMAL,<0,2.1944,-0.3618>,<-1,0,0>,<0.0011,2.1942,-0.3618>,USE THEO=YES
ENDMEAS/
MOVE/POINT,NORMAL,<-0.4913,2.1944,1.5459>
A2 =ALIGNMENT/START,RECALL:STARTUP,LIST=YES
ALIGNMENT/LEVEL,ZPLUS,PLN2
ALIGNMENT/TRANS,ZAXIS,PLN2
ALIGNMENT/ROTATE,XPLUS,TO,LIN2,ABOUT,ZPLUS
ALIGNMENT/TRANS,YAXIS,LIN2
ALIGNMENT/TRANS,XAXIS,PNT2
ALIGNMENT/END
V1 =LOOP/START,ID=YES,NUMBER=2,START=1,SKIP=,
OFFSET:XAXIS=0,YAXIS=-11.5,ZAXIS=0,ANGLE=0
MOVE/POINT,NORMAL,<3.9146,2.1944,2.4398>
PLN3-DATUM A =FEAT/PLANE,CARTESIAN,TRIANGLE
THEO/<5.9086,2.2725,0>,<0,0,1>
ACTL/<5.9087,2.2722,0.0033>,<0.0004246,-0.0001787,0.9999999>
MEAS/PLANE,7
HIT/BASIC,NORMAL,<4.3281,3.3546,0>,<0,0,1>,<4.3283,3.3 544,0.0044>,USE THEO=YES
HIT/BASIC,NORMAL,<4.0306,1.4526,0>,<0,0,1>,<4.0305,1.4 522,0.0041>,USE THEO=YES
HIT/BASIC,NORMAL,<5.7436,3.0695,0>,<0,0,1>,<5.7437,3.0 693,0.0031>,USE THEO=YES
HIT/BASIC,NORMAL,<5.8434,1.4538,0>,<0,0,1>,<5.8436,1.4 534,0.0029>,USE THEO=YES
HIT/BASIC,NORMAL,<7.2594,3.4126,0>,<0,0,1>,<7.2596,3.4 124,0.0031>,USE THEO=YES
HIT/BASIC,NORMAL,<6.8504,2.1411,0>,<0,0,1>,<6.8502,2.1 408,0.0029>,USE THEO=YES
HIT/BASIC,NORMAL,<7.3046,1.0234,0>,<0,0,1>,<7.3047,1.0 23,0.0027>,USE THEO=YES
ENDMEAS/
MOVE/POINT,NORMAL,<7.3046,0.3131,1.0796>
MOVE/POINT,NORMAL,<7.3046,-0.5902,-0.3132>
PLN4-DATUM B =FEAT/PLANE,CARTESIAN,TRIANGLE
THEO/<4.6859,0,-0.7652>,<0,-1,0>
ACTL/<4.6858,0.0045,-0.7651>,<0.0000139,-0.9999988,-0.0015266>
MEAS/PLANE,8
HIT/BASIC,NORMAL,<9.0292,0,-0.683>,<0,-1,0>,<9.0294,0.0044,-0.683>,USE THEO=YES
(CODE CUT TO FIT)
ENDMEAS/
MOVE/POINT,NORMAL,<0.1156,-0.5425,2.3897>
MOVE/POINT,NORMAL,<0.1156,4.8792,2.1768>
MOVE/POINT,NORMAL,<10.7453,4.8792,2.63>
MOVE/POINT,NORMAL,<10.7453,5.1562,-0.4423>
PLN5-BACKSIDE COMPARATIVE =FEAT/PLANE,CARTESIAN,TRIANGLE
THEO/<6.3431,4.59,-0.7871>,<0,1,0>
ACTL/<6.343,4.5968,-0.7869>,<-0.0000426,0.9999986,-0.0016886>
MEAS/PLANE,8
HIT/BASIC,NORMAL,<11.4177,4.59,-0.638>,<0,1,0>,<11.4179,4.5971,-0.6379>,USE THEO=YES
(CODE CUT TO FIT)
ENDMEAS/
MOVE/POINT,NORMAL,<0.1326,4.774,0.2651>
MOVE/POINT,NORMAL,<-0.331,4.774,0.0903>
PLN6-DATUM C =FEAT/PLANE,CARTESIAN,TRIANGLE
THEO/<0,2.3748,-0.5079>,<-1,0,0>
ACTL/<0.011,2.3744,-0.508>,<-0.9999993,0.0000425,0.0011476>
MEAS/PLANE,9
HIT/BASIC,NORMAL,<0,4.5184,-0.7714>,<-1,0,0>,<0.0103,4.518,-0.7715>,USE THEO=YES
(CODE CUT TO FIT/MEASURING A PLANE)
ENDMEAS/
MOVE/POINT,NORMAL,<-0.7644,0.0617,1.8962>
MOVE/POINT,NORMAL,<2.0946,0.0617,2.4491>
A3 =ALIGNMENT/START,RECALL:STARTUP,LIST=YES
ALIGNMENT/LEVEL,ZPLUS,PLN3-DATUM A
ALIGNMENT/TRANS,ZAXIS,PLN3-DATUM A
ALIGNMENT/ROTATE,YMINUS,TO,PLN4-DATUM B,ABOUT,ZPLUS
ALIGNMENT/TRANS,YAXIS,PLN4-DATUM B
ALIGNMENT/TRANS,XAXIS,PLN6-DATUM C
ALIGNMENT/END
A4 =ALIGNMENT/START,RECALL:STARTUP,LIST=YES
ALIGNMENT/LEVEL,ZPLUS,PLN3-DATUM A
ALIGNMENT/TRANS,ZAXIS,PLN3-DATUM A
ALIGNMENT/ROTATE,YMINUS,TO,PLN4-DATUM B,ABOUT,ZPLUS
ALIGNMENT/TRANS,YAXIS,PLN4-DATUM B
ALIGNMENT/TRANS,XAXIS,PLN6-DATUM C
ALIGNMENT/END
RECALL/ALIGNMENT,INTERNAL,A3
CIR1 =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
THEO/<2.27,13.795,0.4>,<0,0,1>,2.62
ACTL/<2.272,13.7962,0.3979>,<0,0,1>,2.6191
TARG/<2.27,13.795,0.4>,<0,0,1>
START ANG=0,END ANG=361.2
ANGLE VEC=<1,0,0>
DIRECTION=CCW
SHOW FEATURE PARAMETERS=NO
SHOW CONTACT PARAMETERS=YES
NUMHITS=8,DEPTH=0.065,PITCH=0
SAMPLE METHOD=SAMPLE_HITS
SAMPLE HITS=0,SPACER=0
AVOIDANCE MOVE=BOTH,DISTANCE=1
FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
SHOW HITS=NO
CIR2 =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
THEO/<9.235,13.795,0.4>,<0,0,1>,2.62
ACTL/<9.235,13.7961,0.3979>,<0,0,1>,2.6192
TARG/<9.235,13.795,0.4>,<0,0,1>
START ANG=0,END ANG=360
ANGLE VEC=<1,0,0>
DIRECTION=CCW
SHOW FEATURE PARAMETERS=NO
SHOW CONTACT PARAMETERS=YES
NUMHITS=8,DEPTH=0.065,PITCH=0
SAMPLE METHOD=SAMPLE_HITS
SAMPLE HITS=0,SPACER=0
AVOIDANCE MOVE=BOTH,DISTANCE=1
FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
SHOW HITS=NO
MOVE/INCREMENT,<0,0,2>
LOOP/END
MOVE/INCREMENT,<6,0,0>
MOVE/INCREMENT,<0,0,4>
MOVE/POINT,NORMAL,<15.241,-20.3018,9.4919>
LOADPROBE/SUXING PROBE 1MM X 40MM
TIP/T1A0B0, SHANKIJK=-0.0007, 0.0006, 1, ANGLE=0.0216
AUTOCALIBRATE/PROBE, PARAMETER_SET=SUXING PROBE 1MMx40MM 0-0, QUALTOOL_MOVED=NO,
CHECK COLLISION=NO, SHOW_SUMMARY=NO, OVERWRITE_RESULTSFILE=NO
MOVE/INCREMENT,<0,0,2>
TOUCHSPEED/ 1
MOVE/POINT,NORMAL,<0.9019,-1.9748,7.5298>
MOVE/POINT,NORMAL,<0.49,0.3912,3.3834>
MOVE/POINT,NORMAL,<6.9268,2.6651,2.9921>
V2 =LOOP/START,ID=YES,NUMBER=2,START=1,SKIP=,
OFFSET:XAXIS=0,YAXIS=-11.5,ZAXIS=0,ANGLE=0
CIR3 =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
THEO/<7.0004,14.8936,0>,<0,0,1>,0.1005
ACTL/<7.0016,14.8947,0>,<0,0,1>,0.1008
TARG/<7.0004,14.8936,0>,<0,0,1>
START ANG=0,END ANG=360
ANGLE VEC=<1,0,0>
DIRECTION=CCW
SHOW FEATURE PARAMETERS=NO
SHOW CONTACT PARAMETERS=YES
NUMHITS=8,DEPTH=0.04,PITCH=0
SAMPLE METHOD=SAMPLE_HITS
SAMPLE HITS=0,SPACER=0
AVOIDANCE MOVE=BOTH,DISTANCE=1
FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
SHOW HITS=NO
CIR4 =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
THEO/<7.7076,14.8936,0>,<0,0,1>,0.1005
ACTL/<7.7077,14.8953,0>,<0,0,1>,0.1008
TARG/<7.7076,14.8936,0>,<0,0,1>
START ANG=0,END ANG=360
ANGLE VEC=<1,0,0>
DIRECTION=CCW
SHOW FEATURE PARAMETERS=NO
SHOW CONTACT PARAMETERS=YES
NUMHITS=8,DEPTH=0.04,PITCH=0
SAMPLE METHOD=SAMPLE_HITS
SAMPLE HITS=0,SPACER=0
AVOIDANCE MOVE=BOTH,DISTANCE=1
FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
SHOW HITS=NO
CIR5 =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
THEO/<7.354,14.54,0>,<0,0,1>,0.283
ACTL/<7.355,14.5412,0>,<0,0,1>,0.2843
TARG/<7.354,14.54,0>,<0,0,1>
START ANG=0,END ANG=360
ANGLE VEC=<1,0,0>
DIRECTION=CCW
SHOW FEATURE PARAMETERS=NO
SHOW CONTACT PARAMETERS=YES
NUMHITS=8,DEPTH=0.04,PITCH=0
SAMPLE METHOD=SAMPLE_HITS
SAMPLE HITS=0,SPACER=0
AVOIDANCE MOVE=BOTH,DISTANCE=1
FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
SHOW HITS=NO
CIR6 =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
THEO/<7.354,13.04,0>,<0,0,1>,0.283


(TRIMMED SOME CODE AT END)
  • I don't think ground pins are your problem. I have a nested loop program where I can run anywhere from 2 - 48 parts depending on operator input. I use 1/4-20 socket head screws for my pins and it runs just fine. If you would:

    Manual align
    loop start
    dcc align(recalling manual align)
    measure with probe 1
    change probe
    measure with probe 2
    dimension and report
    loop end

    I feel the problems you are having will go away.

    Side note, I don't think you need all those calibrations either

    Calibrate your master probe (yes sphere has moved)
    Calibrate your other probes (no sphere has not moved)
    if this is done before running the program it shouldn't need to be done again until you crash or break thread... Those magnetic connections are fairly repeatable
  • its pretty simple. i use exclusively readpoint alignments in all my programs.
    so i'll write a program, setup my part on my table/fixture system, then at the end of the program do a movepoint to the next part where my first start/readpoint was, then do a recall STARTUP ALIGNMENT (which kinda basically start the program over again), then i'll copy and paste the program x times.

    if i have 5 parts on the table the program runs basically 5 different parts. i group each program together and close it.
    this allows me to stop the program anytime i want and look at data, switch out parts, etc. If i remember correctly looping becomes a mess if you stop in the middle of execution.

    the only disadvantage to my method is that naming becomes a bit of a headache, so you need to plan ahead on how you name features and dimensions so that when you copy/pattern it copies everything and you don't lose track of which part is which.

    i'm simply not smart enough or have enough patience to figure out looping.
  • I am going to try what you just posted. I can 100% verify that I did not recall a single alignment after the startup. I def need to read up on it more because I want the *theory* behind it etched into my skull. Simply clicking and following what the lab people told me to do is frustrating. Thank you! The amount of misinformation/theories/bro-science from everyone including Hexagon is maddening. I've heard so many things from multiple people I'm surprised I haven't lost it yet.
  • ahhh, just learned something invaluable!!! So...pc-dmis DOES automatically make the next alignment you add recall the previous one...but...I'm using auto align! Which for some odd reason calls up startup. I use auto-align almost exclusively on simple squarish parts....so..that's why the program sucks! (or at least part of the reason). If I do auto-alignment and recall the previous alignment and hit okay, it does work though. I thought PC-DMIS was gonna stroke out and blue screen me haha.
  • Whoa there fella! People are trying to help.

    The thing is, (and I'm not meaning to be condescending here) it sounds a little like you may be trying to run before you can walk.

    PCDMIS is incredibly powerful, but also incredibly complex. Your asking for advice on how to achieve something fairly complex, that while doable isn't particularly easy to explain to a relative beginner, and short of someone simply doing it for you will be open to misinterpretation (especially when several people are shouting different ideas at you at once).

    Also what might be confusing, it that there are typically several ways of achieving the same thing (theories/misinformation/bro science).
  • ??? I've been very very appreciative of any help that I can get. If you read any of my comments and felt like I was attacking any of the people answering my questions, I am sorry both to you (and whoever felt like that if they did). If you mean I'm just posting too much/ not giving you enough time to respond I'm sorry. I'm just under somewhat of a time crunch here and I have to figure this stuff out like 6 months ago. Lol, yeah, my life has always been run before you walk. It's just the nature of where I work though, I don't have much of a choice. It's very much a diy figure it out yourself or it's your fault kind of job. I don't mind people throwing different ideas at me because I'll def use all of them in the future (we do so many different setups I've gotta learn all the tricks.) I just need to find a way to make it work for me then modify it. I'm not expecting anybody to "do it" for me either. I've been reading these forums for a year or maybe 2 now, just never posted. It's like learning a new language, you can read so much before you have to converse.
  • Fair enough, I didn't mean to cause offence either.

    I ask again however... Why the need to avoid the probe change mid program? There are loads of looping examples like this which should be easy enough to replicate. Perhaps get a simple liked program (including probe changes) sorted first, then look to improve it?

    Or are you having to manually change and recalibrate the probe? (As I mentioned, I've seen your posts but not had chance to delve into them)
  • I would use this way too. Since all measurement needs reports, it is hard to report those 6 or 4 parts different features in loop with identical IDs.
  • oh okay, I just realized all these comments I had been replying to your original post, sorry bout that. (might have conveyed I was ignoring you) no no no, I'm def not like that. Tbh if I had enough time I would open a youtube channel dedicated to everything I know about PC-DMIS so people wouldn't have to pay for the classes lol.

    Okay, to answer your question: The reason I want to change probes while inspecting parts is for simple efficiency. My background is in mills and lathes. Lets say I was cutting 10 identical parts on a mill and they are all aligned in the X-axis so the machine just has to go over in x to cut it. Instead of drilling 3 holes, and then changing to a tap to drill those 3 holes...I drill 3 holes in ALL 10 parts first (30 holes total), then come back and tap all 30 holes. It's the same with the CMM. I don't have a probe rack (and maybe that's why more of you experienced guys are wondering wth I'm doing), so for me to just push go and walk away it's much more advantageous for the machine to use probe 1 to probe all the features on parts 1, 2, 3, etc...THEN do the tool change and come back and probe all the parts in order again 1, 2, 3. I don't know if I'm explaining myself very well. :'( My social skills are very subpar.
  • Here's a direct answer to your problem with the loop.
    Run it offline with the qty set to 1.
    Save it.
    Then, make the following change:
    You NEED to repeat your initial DCC alignment at the beginning of the loop. then you need to recall previous for all alignments within the loop, AFTER that repeat DCC alignment.

    Once you have the redundant DCC alignment set, (you might want to increase prehit/retract and search distances as well for these initial dcc features)
    Then toggle your loop count back to 2 and save a copy of the routine.
    Set new routine to read-only in explorer browser.
    execute it.
    see what happens & report back.


    STARTUP =ALIGNMENT/START,RECALL:USE_PART_SETUP,LIST=YES
    ALIGNMENT/END
    MODE/MANUAL
    FORMAT/TEXT,OPTIONS, ,HEADINGS,SYMBOLS, ;NOM,TOL,MEAS,DEV,OUTTOL, ,
    LOADPROBE/4MM X 20MM
    TIP/T1A0B0, SHANKIJK=0, 0, 1, ANGLE=0
    PLN1 =FEAT/PLANE,CARTESIAN,TRIANGLE
    THEO/<3.4829,-0.1456,0>,<0,0,1>
    ACTL/<12.1376,32.7697,-26.1905>,<0.0005222,-0.0003643,0.9999998>
    MEAS/PLANE,3
    HIT/BASIC,NORMAL,<1.9889,-0.5175,0>,<0,0,1>,<10.5594,32.265,-26.1899>,USE THEO=YES
    HIT/BASIC,NORMAL,<4.0258,0.8213,0>,<0,0,1>,<12.1295,33 .9516,-26.1901>,USE THEO=YES
    HIT/BASIC,NORMAL,<4.434,-0.7405,0>,<0,0,1>,<13.7238,32.0923,-26.1916>,USE THEO=YES
    ENDMEAS/
    LIN1 =FEAT/LINE,CARTESIAN,UNBOUNDED
    THEO/<-2.1728,-2.295,-0.6305>,<1,0,0>
    ACTL/<6.7011,30.8644,-26.6467>,<1,-0.0003111,0>
    MEAS/LINE,2,ZPLUS
    HIT/BASIC,NORMAL,<-2.1728,-2.295,-0.629>,<0,-1,0>,<6.7011,30.8644,-26.6468>,USE THEO=YES
    HIT/BASIC,NORMAL,<6.84,-2.295,-0.632>,<0,-1,0>,<15.6711,30.8616,-26.6467>,USE THEO=YES
    ENDMEAS/
    PNT1 =FEAT/POINT,CARTESIAN
    THEO/<-2.275,0.0718,-0.3763>,<-1,0,0>
    ACTL/<6.3863,31.8955,-26.567>,<-0.9999996,-0.0009195,0>
    MEAS/POINT,1,WORKPLANE
    HIT/BASIC,NORMAL,<-2.275,0.0718,-0.3763>,<-1,0,0>,<6.3863,31.8955,-26.567>,USE THEO=YES
    ENDMEAS/
    A1 =ALIGNMENT/START,RECALL:STARTUP,LIST=YES
    ALIGNMENT/LEVEL,ZPLUS,PLN1
    ALIGNMENT/TRANS,ZAXIS,PLN1
    ALIGNMENT/ROTATE,XPLUS,TO,LIN1,ABOUT,ZPLUS
    ALIGNMENT/TRANS,YAXIS,LIN1
    ALIGNMENT/TRANS,XAXIS,PNT1
    ALIGNMENT/END
    MODE/DCC
    FLY/ON
    MOVESPEED/ 250
    MOVE/POINT,NORMAL,<4.5328,1.9318,1.0961>
    PLN2 =FEAT/PLANE,CARTESIAN,TRIANGLE
    THEO/<5.7549,2.0759,0>,<0,0,1>
    ACTL/<5.755,2.0756,-0.0033>,<0.0000146,-0.0000177,1>
    MEAS/PLANE,3
    MOVE/POINT,NORMAL,<7.0084,1.6156,0.8147>
    (CODE CUT TO FIT, JUST MOVE POINTS)
    HIT/BASIC,NORMAL,<4.2265,1.4394,0>,<0,0,1>,<4.2264,1.4 389,-0.0033>,USE THEO=YES
    HIT/BASIC,NORMAL,<6.0296,3.0913,0>,<0,0,1>,<6.0298,3.0 911,-0.0033>,USE THEO=YES
    HIT/BASIC,NORMAL,<7.0084,1.6972,0>,<0,0,1>,<7.0086,1.6 967,-0.0033>,USE THEO=YES
    ENDMEAS/
    MOVE/POINT,NORMAL,<6.9272,2.3614,1.4628>
    MOVE/POINT,NORMAL,<6.8272,-0.616,0.8246>
    MOVE/POINT,NORMAL,<0.1955,-0.616,1.1066>
    MOVE/POINT,NORMAL,<0.1955,-0.3262,-0.6042>
    LIN2 =FEAT/LINE,CARTESIAN,UNBOUNDED
    THEO/<0.2769,0,-0.6506>,<1,0,0>
    ACTL/<0.2769,0.0045,-0.6505>,<1,-0.0000029,0>
    MEAS/LINE,2,ZPLUS
    HIT/BASIC,NORMAL,<0.2769,0,-0.6573>,<0,-1,0>,<0.2769,0.0045,-0.6573>,USE THEO=YES
    HIT/BASIC,NORMAL,<8.968,0,-0.6438>,<0,-1,0>,<8.9682,0.0045,-0.6437>,USE THEO=YES
    ENDMEAS/
    MOVE/POINT,NORMAL,<8.968,-0.4155,1.3796>
    MOVE/POINT,NORMAL,<-0.591,-0.4155,1.5577>
    MOVE/POINT,NORMAL,<-0.6887,-0.4155,-0.3356>
    PNT2 =FEAT/POINT,CARTESIAN
    THEO/<0,2.1944,-0.3618>,<-1,0,0>
    ACTL/<0.0011,2.1942,-0.3618>,<-1,0,0>
    MEAS/POINT,1,WORKPLANE
    HIT/BASIC,NORMAL,<0,2.1944,-0.3618>,<-1,0,0>,<0.0011,2.1942,-0.3618>,USE THEO=YES
    ENDMEAS/
    MOVE/POINT,NORMAL,<-0.4913,2.1944,1.5459>
    A2 =ALIGNMENT/START,RECALL:STARTUP,LIST=YES
    ALIGNMENT/LEVEL,ZPLUS,PLN2
    ALIGNMENT/TRANS,ZAXIS,PLN2
    ALIGNMENT/ROTATE,XPLUS,TO,LIN2,ABOUT,ZPLUS
    ALIGNMENT/TRANS,YAXIS,LIN2
    ALIGNMENT/TRANS,XAXIS,PNT2
    ALIGNMENT/END
    V1 =LOOP/START,ID=YES,NUMBER=2,START=1,SKIP=,
    OFFSET:XAXIS=0,YAXIS=-11.5,ZAXIS=0,ANGLE=0
    
    <<< I Literally copied and pasted your initial alignment features from above in here. and you can do the same (the demon will paramatrically update all the coords, but you will need to rename feature id's to be unique).>>>>>
    
    MOVE/POINT,NORMAL,<4.5328,1.9318,1.0961>
    PLN2 =FEAT/PLANE,CARTESIAN,TRIANGLE
    THEO/<5.7549,2.0759,0>,<0,0,1>
    ACTL/<5.755,2.0756,-0.0033>,<0.0000146,-0.0000177,1>
    MEAS/PLANE,3
    MOVE/POINT,NORMAL,<7.0084,1.6156,0.8147>
    (CODE CUT TO FIT, JUST MOVE POINTS)
    HIT/BASIC,NORMAL,<4.2265,1.4394,0>,<0,0,1>,<4.2264,1.4 389,-0.0033>,USE THEO=YES
    HIT/BASIC,NORMAL,<6.0296,3.0913,0>,<0,0,1>,<6.0298,3.0 911,-0.0033>,USE THEO=YES
    HIT/BASIC,NORMAL,<7.0084,1.6972,0>,<0,0,1>,<7.0086,1.6 967,-0.0033>,USE THEO=YES
    ENDMEAS/
    MOVE/POINT,NORMAL,<6.9272,2.3614,1.4628>
    MOVE/POINT,NORMAL,<6.8272,-0.616,0.8246>
    MOVE/POINT,NORMAL,<0.1955,-0.616,1.1066>
    MOVE/POINT,NORMAL,<0.1955,-0.3262,-0.6042>
    LIN2 =FEAT/LINE,CARTESIAN,UNBOUNDED
    THEO/<0.2769,0,-0.6506>,<1,0,0>
    ACTL/<0.2769,0.0045,-0.6505>,<1,-0.0000029,0>
    MEAS/LINE,2,ZPLUS
    HIT/BASIC,NORMAL,<0.2769,0,-0.6573>,<0,-1,0>,<0.2769,0.0045,-0.6573>,USE THEO=YES
    HIT/BASIC,NORMAL,<8.968,0,-0.6438>,<0,-1,0>,<8.9682,0.0045,-0.6437>,USE THEO=YES
    ENDMEAS/
    MOVE/POINT,NORMAL,<8.968,-0.4155,1.3796>
    MOVE/POINT,NORMAL,<-0.591,-0.4155,1.5577>
    MOVE/POINT,NORMAL,<-0.6887,-0.4155,-0.3356>
    PNT2 =FEAT/POINT,CARTESIAN
    THEO/<0,2.1944,-0.3618>,<-1,0,0>
    ACTL/<0.0011,2.1942,-0.3618>,<-1,0,0>
    MEAS/POINT,1,WORKPLANE
    HIT/BASIC,NORMAL,<0,2.1944,-0.3618>,<-1,0,0>,<0.0011,2.1942,-0.3618>,USE THEO=YES
    ENDMEAS/
    MOVE/POINT,NORMAL,<-0.4913,2.1944,1.5459>
    A2 =ALIGNMENT/START,RECALL:STARTUP,LIST=YES
    ALIGNMENT/LEVEL,ZPLUS,PLN2
    ALIGNMENT/TRANS,ZAXIS,PLN2
    ALIGNMENT/ROTATE,XPLUS,TO,LIN2,ABOUT,ZPLUS
    ALIGNMENT/TRANS,YAXIS,LIN2
    ALIGNMENT/TRANS,XAXIS,PNT2
    ALIGNMENT/END
    ​
    <<< I Literally copied and pasted your initial alignment features from above in here. and you can do the same (the demon will paramatrically update all the coords, but you will need to rename feature id's to be unique).>>>>>
    <<<From here on out, Recall/USE PREVIOUS ALIGNMENT for all other aligns.>>>
    
    MOVE/POINT,NORMAL,<3.9146,2.1944,2.4398>
    PLN3-DATUM A =FEAT/PLANE,CARTESIAN,TRIANGLE
    THEO/<5.9086,2.2725,0>,<0,0,1>
    ACTL/<5.9087,2.2722,0.0033>,<0.0004246,-0.0001787,0.9999999>
    MEAS/PLANE,7
    HIT/BASIC,NORMAL,<4.3281,3.3546,0>,<0,0,1>,<4.3283,3.3 544,0.0044>,USE THEO=YES
    HIT/BASIC,NORMAL,<4.0306,1.4526,0>,<0,0,1>,<4.0305,1.4 522,0.0041>,USE THEO=YES
    HIT/BASIC,NORMAL,<5.7436,3.0695,0>,<0,0,1>,<5.7437,3.0 693,0.0031>,USE THEO=YES
    HIT/BASIC,NORMAL,<5.8434,1.4538,0>,<0,0,1>,<5.8436,1.4 534,0.0029>,USE THEO=YES
    HIT/BASIC,NORMAL,<7.2594,3.4126,0>,<0,0,1>,<7.2596,3.4 124,0.0031>,USE THEO=YES
    HIT/BASIC,NORMAL,<6.8504,2.1411,0>,<0,0,1>,<6.8502,2.1 408,0.0029>,USE THEO=YES
    HIT/BASIC,NORMAL,<7.3046,1.0234,0>,<0,0,1>,<7.3047,1.0 23,0.0027>,USE THEO=YES
    ENDMEAS/
    MOVE/POINT,NORMAL,<7.3046,0.3131,1.0796>
    MOVE/POINT,NORMAL,<7.3046,-0.5902,-0.3132>
    PLN4-DATUM B =FEAT/PLANE,CARTESIAN,TRIANGLE
    THEO/<4.6859,0,-0.7652>,<0,-1,0>
    ACTL/<4.6858,0.0045,-0.7651>,<0.0000139,-0.9999988,-0.0015266>
    MEAS/PLANE,8
    HIT/BASIC,NORMAL,<9.0292,0,-0.683>,<0,-1,0>,<9.0294,0.0044,-0.683>,USE THEO=YES
    (CODE CUT TO FIT)​