Coupe of observations/questions:
You say the rad has been verified at 2.75 by two other methods - what methods are these? Any system that calculates the radius from points would be susceptible to the same issue (this is not a PC-Dmis issue, it's a math issue).
The point of using the fixed rad method is that, by calculating a circle from points over an insufficient degree of arc, can give an inaccurate and non-repeatable centre point, giving bogus radius values.
So 1rst method is giving it to an external metrology laboratory. It came up at 2.75 something. Second method is a mixture of results such as: Isomat projector (2d visual measurement), CMM simple measurement with least squares and finally testing fixed dia pins and seeing how they best fit on the rad cutoff.
To tell you the truth, I am disappointed of this method because it is dependent on the value the fxd rad we initially give. As said, I tried 2.8 (nominal) and got repeatable close to 2.8 results and then I tried using...
CYL_2_8.D (a least square circle that was taken before that gives 2.75..mm) as the nominal D and the results I was getting were repeatable 2.75 something.
The repeatability is there but not the accuracy.
Normally when i use this method i use the fixed radius and then change back to the original alignment and report out T values of each point. It doesnt give a radial value but it gives nominal surface deviation from the intended surface.