Calibration

Hi, So when calibrating probe it is like I'm 10mm off.I calibrate my probe at A0,B0 ok but then it moves to A90,B-90 its hitting the calibration sphere.So I tried to start again manul first touch but still happened. I then tried did manul touch A90,B-90 but when it went to do the A0,B0 it was off by about 10mm.Please any thoughts
Parents
  • Could be a corrupted prb file. Print screen your probe build and delete the .prb file then rebuild the probe.

    Also make sure your probe head is physically oriented correctly. I align mine to take a point at A90B0 then taking the same point at A90B180. PITA but only have to do it once.