Hi, So when calibrating probe it is like I'm 10mm off.I calibrate my probe at A0,B0 ok but then it moves to A90,B-90 its hitting the calibration sphere.So I tried to start again manul first touch but still happened. I then tried did manul touch A90,B-90 but when it went to do the A0,B0 it was off by about 10mm.Please any thoughts
Could be a corrupted prb file. Print screen your probe build and delete the .prb file then rebuild the probe.
Also make sure your probe head is physically oriented correctly. I align mine to take a point at A90B0 then taking the same point at A90B180. PITA but only have to do it once.
Could be a corrupted prb file. Print screen your probe build and delete the .prb file then rebuild the probe.
Also make sure your probe head is physically oriented correctly. I align mine to take a point at A90B0 then taking the same point at A90B180. PITA but only have to do it once.
Sora can you please expand on the topic of making sure that your probe head is Physically orientated correctly. How do you verify this ? What do you take points on ? the table ?