A bit over three and a half years ago we bought our first continuous-scanning-capable CMM, a Global S Blue 9.12.8 with a HP-S-X1 scanning head (I believe that's just Hexagon's rebranding of the Leitz LSPX line) and Hexagon's 7.5 degree indexable auto wrist, the HH-AS8-T7.5. What the machine has allowed us to do has been amazing, and on the whole we're very happy with the system.
We've never crashed it badly (as in, nothing big enough to make any visible dents on either the scanning probe or the wrist), though I did put too much torque on the scanning head trying to set up a star probe at one point and had to replace that.
Over the course of the last week or so we've had repeated instances of the wrist not seating - sometimes it will try multiple times and then succeed, other times it will not succeed and then just give up. This morning I had to power cycle the entire system to get it to work after it gave up. From what I've seen this behavior points to a failure of the wrist that will require a replacement.
We use the machine frequently over two shifts, five days a week, but beyond the regular use I don't think we do anything that should be particularly hard on it. I see pictures in Hexagon's marketing literature of these heads waving around huge laser scanners that must weigh a good bit more than the scanning head we have on it. Is this seemingly-short lifespan normal? Did we just get a bad egg? Is there something in our use or programming that could be causing more than typical wear? My boss is not very happy at the thought of having to replace these so frequently. Any ideas would be appreciated.
One of our heads started doing the same after 8 years service Hexagon tech guy said they can start drifting with time so re-phased the head to correct, lasted another 3 years.
That misalignment definitely fits with some of the behavior we've seen; we've also had some odd accuracy issues with it too, though, so this time I'm thinking it would probably be best to go through with the replacement.
Well, at this point we're almost finished finalizing the RBE with the parts department. I'm confident we can change it ourselves. We wanted to avoid paying $1000 on top of the replacement for someone to come out and say, "Yep, it's broke" and take the machine down for longer still.
Also, is there a way to change the overall tool change speed? I know there's the "Docking Speed" in the Probe Changer calibration dialog, but that only seems to take effect when it's holding a tip - ours go in to pick up tips at the program speed, which has occasionally been pretty zippy.
TESA heads are nothing compared to Renishaw heads for being robust. We go thru a lot of heads in a year but... our CMMs run 24/6 and some run 24/7. I never saw the issue when all of our older CMMs were running PH9s and PH10s.
Four years on a Hexagon / Tesa head with constant use? Probably not out of the question. We've had Tesas' forever, too much use wears the internal gears. The newer Hexagon heads seem to be much better. At least for our money.