I've been having issues with the probe rack positions changing after shutting down and rehoming the machine. The rack shifts about 1-2mm from where it was calibrated. I've been looking for the tool.dat file but I can't find it. I'm running 2018 R2. I read that it can be an issue with the tool.dat file or the machine's home changes and will throw off the position every time the machine is homed.
For the changer, it's the toolc.dat (tool.dat is about artifacts)
The homing repeatability depends on the cmm, but 1 - 2 mm is much too large.
Maybe the probe was not calibrated when you performed the rack calibration ?