Tips Calibrated, wrist angles way off?

I have a program that uses T1A0B0, T1A90B0, T1A-90B0 and T1A90B180 and I tried to measure from the front of the part half a cylinder, then rotating to -90B180 and measuring the back half to construct a center axis, but the numbers were way off.

So then I went and re-calibrated and grabbed a gage block and made the attached program. I measured the front plane of the gage block and it related the A0B0 and A-90B0 correctly, but the plane on the back of the gage block using -90B180. It measures Y=76.802 instead of Y=76.200 so off by 0.6mm on the rear plane. Then I did the front plane with A90B180 and it was measuring at -0.3mm. I calibrated all tips and the probe results, are added. Did I forget to calibrate something or is something off with my wrist? Anyone have ideas as to what is going on or what I could have missed?

Attached Files
  • How did you calibrate your probes? Do you have a master? Did you calibrate your master first?
  • Calibrated A0B0 man+DCC then calibrated the rest with a didn't move command. I don't think I set a master, I'll look at how to define and calibrate a master, I don't remember how to do that, it's been a couple years since I programmed, still trying to "get back on the bike".
  • Calibrating master probe is easy. Choose a probe with a short shank preferably 3mmx20mm, when you hit measure tab select YES & take one hit on top of sphere. Then for the rest of probes choose NO. And make sure calibration touch speeds are the same as in the programs.
  • As Matt stated important to do the master probe first, always, also you're Prb Rdv results look out too Disappointed
  • There is something wrong in the hit data after the first alignment the Y axis is not showing the correct nominal, also do not use the minus symbol in names use underscore.
  • I have a program that uses T1A0B0, T1A90B0, T1A-90B0 and T1A90B180 and I tried to measure from the front of the part half a cylinder, then rotating to -90B180 and measuring the back half to construct a center axis, but the numbers were way off.

    So then I went and re-calibrated and grabbed a gage block and made the attached program. I measured the front plane of the gage block and it related the A0B0 and A-90B0 correctly, but the plane on the back of the gage block using -90B180. It measures Y=76.802 instead of Y=76.200 so off by 0.6mm on the rear plane. Then I did the front plane with A90B180 and it was measuring at -0.3mm. I calibrated all tips and the probe results, are added. Did I forget to calibrate something or is something off with my wrist? Anyone have ideas as to what is going on or what I could have missed?


    Whenever the needle is bent, you always get those kind of data from my experience.
  • Calibrated A0B0 man+DCC then calibrated the rest with a didn't move command. I don't think I set a master, I'll look at how to define and calibrate a master, I don't remember how to do that, it's been a couple years since I programmed, still trying to "get back on the bike".


    For testing purposes, I would calibrate A0B0 answering yes the sphere has moved to either man or dcc and then calibrate the rest of the tip angles answering NO for each.