I am using a 0.7X20mm probe (long thin shaft). What are the best calibration settings for this probe? It keeps failing calibration for a few tip angles (.0004 deviation). I am using DCC+DCC 31 hits (5 levels) with a start angle of 10 and end angle of 90.
This mostly depends on shaft diameter and material, but when things get very flakey, or if you have a thin part: touchspeed 1.0, Max force 0.024N, Low force 0.012N, Upper force 0.02N, Trigger force 0.016N
Yep. We have .5mmx20mm probe that gets qualified at 1% touch speed Renishaw PH10T. At the other CMM that's got scanning head .5x45mm gets qualified at 0.5% speed because we notices significant bending during calibration. It works well now.