4.5.4-SF Wont calibrate 1mmx22mm Probe

Hey guys. We are having an issue with our 4.5.4 where it wont calibrate the 1mm probe. So far I have changed the Tp-20 module, the 1mm probe it's self and even created a new probe profile. Have reset all tips and recalibrated the master probe and the 1.5MM probe just fine. Im out of ideas as to what the issue could be.
Please Help
Parents
  • I had an issue like this, and it turned out to be a damaged qualification artifact. Someone had damaged/dropped it and the ceramic ball broke free from the pilar The ball was just loose enough to allow for a tiny amount of movement when being qualified. Could be something else causing your issue but it is something that can be easily overlooked.Rolling eyes
Reply
  • I had an issue like this, and it turned out to be a damaged qualification artifact. Someone had damaged/dropped it and the ceramic ball broke free from the pilar The ball was just loose enough to allow for a tiny amount of movement when being qualified. Could be something else causing your issue but it is something that can be easily overlooked.Rolling eyes