Your Products have been synced, click here to refresh
.. . I am going out on a limb and assuming the Cal-sphere to disc probe size relation could be the issue as the sphere only .59025" and the only disc we had laying around was an 18mmx7.5mm.
You are right in your assumption - 1.0'' will sort this out. We have both size spheres onsite. I did try on 1.0'' and it works with analogue probe/scan module:
Ø18 disc should clash on .59025" (Ø15mm) sphere as per trigonometry calculation (in millimeters):
Regards
For all whom need to order stuff that you cannot find on the site, email me with the address below and I can send you the catalog I got from support with the cal-sphere I need within.
Thanks all for help, will report back to post when I get the sphere and successfully calibrate this disc.
How large a disc and how large a cal-sphere? That's what did it for me. You can also change that figure inside settings editor some.