PCDMIS 2020 R1 Hexagon SF5-4-5 machine. I use a master probe, a change rack and an automated calibration routine for calibration. The master probe measures one tip and is used to set the location of the of the calibration sphere. Master probe is 3mm x 20 stylus straight. The routine then switches a probe build with a five way with only 2 positions filled one tip pointing to the back of the CMM and one towards the front. It begins to auto calibrate through a series of tips stored in a parameter set. All of the tips with a A0 angle calibrate fine but when it rotates to a A65 or A90 it misses the ball completely. I have 9 other machines running this calibration program just fine. I have remade all the probe files. I have replaced the modules and probe builds for the master and the right angle insuring they are straight etc. Still the same issue but with only this one machine. It is almost like the rotation is not correct on the head. Any ideas what is going on here? Thanks