Thanks for reading. I recently measured a part manually with measured circles and aligned said part. However, EVERYTHING is coming back at nominal. TP is exactly 0, and my diameters are exact. Any ideas? Thanks for your time.
  • Your nominal and measured values will be the same when measuring manually for the first time. Have you gone back and corrected the nominal relationship between features?
  • Thanks for your response Jacob. I greatly appreciate it.

    So, on the intial run through while live programming, the CMM will not relay measurements? And please forgive my ignorance, but what is meant by going back and correcting the nominal relationship between features?

    Thanks for your time.
  • When you manually measure a feature, say a diameter, PC-DMIS doesn't know what the size is supposed to be. It will assume that the measured size is the nominal size. If you don't go back to the feature and correct the nominal value while retaining the measured values, you will see no deviation from nominal.

    Example: I manually probe a hole that is .250 in diameter without previously defining it in PC-DMIS. I measure the hole at .2475 manually. PC-DMIS has no way to know that it should be .250, so it assumes the nominal is .2475. If I don't change the nominal value to .250, I'll see no deviation.

    You have to change the nominal definition (including size where applicable, orientation, and location) of all features to the requirements of the drawing that you're working to.
  • If you have the blue print, you can find theo values, so deviations won't be zeros.