My machine tries to rotate to say a90b80 and it only gets to say a40b80…same issue with a few other tip angles…then, I believe it tries to probe the cal sphere with the head at the wrong angle, and of course the tip is no where near the cal sphere so “probing not within limitation.” I believe that is the error. I haven’t seen it happen in action yet, but when I returned to the inspection lab I believe the red light on the head is turned off and the head is disengaged. If I recall correctly, I don’t think I had to hit the estop to reset the jog box. I cancelled running the program, then I typed tip/T1A0B0. When it asked to rotate to A0B0 I clicked yes and it worked fine. Then I tried the same thing with the tip angle in question (a90b80). Again, rotated fine. So I’m not understanding why my program won’t run through…
Any ideas what the problem might be? I tried changing the probe module and wiping down the scales. No luck.
So far, this issue is only happening with my auto calibrate program (which calibrates “all tip angles”). I haven’t tested running a part because the machine is not fully calibrated. However, my auto calibrate program has been unchanged for about a year now…I don’t believe I have even added a single tip angle recently…my autocalibrate program has been working so well up until now!
Sorry for the wall of text, I am just trying to figure this out. Tomorrow I can try calibrating different probes and see if there’s the same issue with the other probes.
How do I change the wrist rotation speed? I tried changing WRISTSPEED/ from 100 to 1 and it appears to be rotating at the same speed as before the change...
The "Display absolute speeds" box is checked if that helps, and the "Top speed" is set to 300mm/sec. Also "Automatically Adjust Probe Head Wrist" is turned off.
Edit: I did some research and, apparently, it's not possible to adjust the wrist speed on this head (HH-A-T5)...can someone please confirm that?
So I created a program that repeatedly (about 20 times) changes tip angles from A0B0 to the problematic tip angles. No issues. Maybe I can figure out what tip angle is prior to the bad tip angles in my autocal program...then test that. Like A90b60 to A90B80. This is so weird that I can't repeat the issue.
Sorry I'm not sure how old it is. It is our older 7.10.7 machine of the two though. We barely use it but I wouldn't be surprised if it had a lot of abuse over the years. I've been here for a year.
It's Hexagon and not TESA, so it's not to old.
On the photo, the red light seems to be off.
Did you check all the wiring, behind the electronic (ex tesastar-ae) ?
There's a solution to send instruction to the head via a port com (or usb, depending on electronics).
It's described in the user's manual.
Thank you JEFFMAN. I just watched the autocalibrate program run for an hour. What I think I just saw was very weird. It goes from calibrating angle A90B-75 to A90B-60. It did the rotation fine then it randomly rotated to (approximately) A30B-60 and the light turned off...