I've seen this before. Turned out to be that the machine was "constipated" with data between the PC and the controller. This was an a large machine that could very well have seriously injured people. Fortunately, it only caused 6-7 probe heads and 2 cars to be destroyed. Took Hexagon almost 2 years to find the cause, although they paid for everything except the cars.
this also happens if we are in the middle of an analog scan and cancel before its completed. the cmm will still keep moving.
neil.challinor It may be the case where defined scans cannot be stopped without pressing E-Stop but that is not what I seen and it doesn't appear to be true for the original poster either.
What I seen was on a Global Performance with a DC240 controller, PH10MQ, TP-20 probe. You are just measuring typical stuff (points, circles, ...) and then you can end up in this situation. Nothing short of a controller reboot or allowing it to complete the buffered commands will stop it and that includes pressing E-Stop. This company described it to me and I actually experienced it once when working on the machine. E-Stop shuts the machine off but once machine start is pressed it continues where it left off.
I am almost certain it is a firmware bug. In PC-DMIS, when execution of anything is cancelled, the software sends a binary sequence (similar to CTRL-C or CTRL-Z) that the controller interprets as a stop and purge. For whatever reason the buffered commands are not purged in some cases and you end up with a zombie CMM.
Recreating this is not easy. As I mentioned the company almost could do it but not quite. There is some detail or step missing that was critical to recreating this problem. This company just accepted that this can happen on this particular machine from time to time and deal with it.