Calibration error diameter exceeds limits

I am using a TP-20. I got a new tip size we havent used before (.7mm x 20mm ). I built a new probe file it is correct except HP-TB-2 is not in the list anymore, so i am using a HP-TBe. I have done this with other tip builds without a problem. When i try to calibrate all the tip angles over half come up with the error probe diameter exceeds limits, i try re-running those angles but get same results any suggestions?
Parents
  • If 's suggestion of slowing touchspeed down doesn't help. check the following:
    1: ensure probe is thoroughly tightened to the module.
    2: clean the ruby and your cal sphere (ck-200 goo or IPA with a lint free wipe)
    3: look under a scope at that specific probe stylus tip. In super rare occasions, i've received probes with excessive glue on the ruby-shank interface, and it affects that stylus's ability to function.
Reply
  • If 's suggestion of slowing touchspeed down doesn't help. check the following:
    1: ensure probe is thoroughly tightened to the module.
    2: clean the ruby and your cal sphere (ck-200 goo or IPA with a lint free wipe)
    3: look under a scope at that specific probe stylus tip. In super rare occasions, i've received probes with excessive glue on the ruby-shank interface, and it affects that stylus's ability to function.