I'm using a Renishaw TTP and I typically use a 4mm ball x 30 or 50mm stylus. I have 3 TP20 modules, SF, MF, and EF. I check steel and aluminum parts. Does it matter which module I use? And should I use different modules for the 30mm and the 50mm? Any recommendations?
You should be able to use any of those modules for a 4 X 30 or a 4 X 50, but the SF will be the most accurate. Check out Renishaw's website for more info. Link below:
Yeah, I thought the SF might be the most accurate. I'm fairly new to the company I'm at, but I think they've been using some incorrect probe builds. Like using a TP20 6W for a 4x50 for example. Could be why I'm getting some questionable results. Thank you very much.
TP20 modules have a lifespan of approximately 200,000 hits.
--Lifespan is much shorter if they are crashed, especially over traveling in Z (compressing the module's internal components).
Definitely anticipate TP20 replacements often, and learn how to spot when they are nearing end of life (false hits, contacting part without detecting hits for example)
Another word of caution about TP20's:
There's standard order "Inhibit" modules which are susceptible to magnetic interference.
You can order NON-INHIBIT modules, which are not affected by magnetic interreference.
Thank you, I swapped out the 6W and EF modules for a MF when using a 50mm stylus and I'm getting much better results. It also had a lot less mileage on it (I think). I'm measuring a part with a 13 micron tolerance, so I needed a lot more accuracy than what I was getting. I think my predecessor may have been using some questionable probe builds for years.
MCR20's "should" be connected to the machine, but you can absolutely use them without any connecting. The Inhibit and Non-Inhibit are physically identical in size and mount interface.