Wrist error. Possible pneumatic issue?

Hello all,

Happy new year! We were having this air pressure issue this morning with our global dea where the air pressure was at around 0.4 MPa. We increased the air pressure to 0.55MPa as per manual and the error disappeared. The CMM makes a homing but stops and gives a wrist error message when trying to change probe/change angle. PH10M Head / TP200 module. Can this 2 issues be related? Does the head engage disengage system is performed pneumatically? Anyone got any ideas?