TP20 Calibration

I have 2 std force modules that have the stylus mounting thread very slightly off center, you can see this by rolling across the Cmm surface table. If a calibration routine is run, will PCDMIS compensate for this?

Parents
  • Calibrating a probe will tell the machine where the probe's spherical centroid is and will work great even if bent a bit... Until you try to measure with the probe deep into a part.

    The machine can't discern or compensate for when you try to measure something deep into a bore, and the probe's shank/shaft/extension is contacting the part before the sphere does (aka a "Shank hit").  This will produce terrible dimensional results in-error. 

    I've dealt with this for TP200's.  Some new, directly out of the box from Renishaw had >1.5° projected axial error.  It doesn't sound like much, but it's enough to shank-hit a 60mm deep 3mm sphere with a 2mm diameter shank.

Reply
  • Calibrating a probe will tell the machine where the probe's spherical centroid is and will work great even if bent a bit... Until you try to measure with the probe deep into a part.

    The machine can't discern or compensate for when you try to measure something deep into a bore, and the probe's shank/shaft/extension is contacting the part before the sphere does (aka a "Shank hit").  This will produce terrible dimensional results in-error. 

    I've dealt with this for TP200's.  Some new, directly out of the box from Renishaw had >1.5° projected axial error.  It doesn't sound like much, but it's enough to shank-hit a 60mm deep 3mm sphere with a 2mm diameter shank.