Need some help with journal file

Trying to process a probe tip calibration journal file, and we keep getting this error.



I am stuck, is this a settings issue?

journal file:

97 15 1 15 I 0059003A005C00500052004F00420045002000500
97 15 052004F004700520041004D0053005C00540045004D00500
97 15 04C0041005400450053005C00500052004F0042004500200
97 15 0430041004C004900420052004100540049004F004E005C0
97 15 056004D0043002000310035005C0056004D0043002300310
97 15 035005F00500052004F00420045002000430041004C00490
98 15 0420052004100540049004F004E002E005000520047
13 15 0 9
7 15 0.0 0.0 0.0
10 15 0 16317 1 .4137 -.0000 6.3091 .0000 0.0
10 15 0 16317 2 .2932 .2932 6.3091 .0000 0.0
10 15 0 16317 3 -.0000 .4144 6.3091 .0000 0.0
10 15 0 16317 4 -.2926 .2926 6.3091 .0000 0.0
10 15 0 16317 5 -.4127 -.0000 6.3091 .0000 0.0
10 15 0 16317 6 -.2912 -.2912 6.3091 .0000 0.0
10 15 0 16317 7 .0000 -.4122 6.3091 .0000 0.0
10 15 0 16317 8 .2919 -.2919 6.3091 .0000 0.0
10 15 0 16317 9 .2962 .0000 6.6052 .0000 0.0
10 15 0 16317 10 .1478 -.2560 6.6047 .0000 0.0
10 15 0 16317 11 -.1477 -.2557 6.6043 .0000 0.0
10 15 0 16317 12 -.2957 .0000 6.6047 .0000 0.0
10 15 0 16317 13 -.1481 .2566 6.6053 .0000 0.0
10 15 0 16317 14 .1484 .2569 6.6057 .0000 0.0
10 15 0 16317 15 -.0000 -.0000 6.7300 .0000 0.0
4 15 0 16317
3 15
9 15
97 15 1 15 I 0059003A005C00500052004F00420045002000500
97 15 052004F004700520041004D0053005C00540045004D00500
97 15 04C0041005400450053005C00500052004F0042004500200
97 15 0430041004C004900420052004100540049004F004E005C0
97 15 056004D0043002000310035005C0056004D0043002300310
97 15 035005F00500052004F00420045002000430041004C00490
98 15 0420052004100540049004F004E002E005000520047
13 15 0 9
7 15 0.0 0.0 0.0
10 15 0 16317 1 .4139 -.0000 6.3091 .0000 0.0
10 15 0 16317 2 .2933 .2933 6.3091 .0000 0.0
10 15 0 16317 3 -.0000 .4143 6.3091 .0000 0.0
10 15 0 16317 4 -.2925 .2925 6.3091 .0000 0.0
10 15 0 16317 5 -.4127 -.0000 6.3091 .0000 0.0
10 15 0 16317 6 -.2911 -.2911 6.3091 .0000 0.0
10 15 0 16317 7 .0000 -.4121 6.3091 .0000 0.0
10 15 0 16317 8 .2920 -.2920 6.3091 .0000 0.0
10 15 0 16317 9 .2962 .0000 6.6052 .0000 0.0
10 15 0 16317 10 .1478 -.2560 6.6047 .0000 0.0
10 15 0 16317 11 -.1476 -.2557 6.6043 .0000 0.0
10 15 0 16317 12 -.2956 .0000 6.6046 .0000 0.0
10 15 0 16317 13 -.1481 .2565 6.6052 .0000 0.0
10 15 0 16317 14 .1483 .2569 6.6057 .0000 0.0
10 15 0 16317 15 -.0000 -.0000 6.7299 .0000 0.0
4 15 0 16317
3 15
9 15


error log:

C:\ProgramData\WAI\PC-DMIS\Journal\DPRNT.OUT
Monday, May 15, 2017 07:41
1. Machine ID Changed in Journal File.
2. Point (1) was Not Found in Journal File for Feature SPH_TIP1
3. Point (2) was Not Found in Journal File for Feature SPH_TIP1
4. Point (3) was Not Found in Journal File for Feature SPH_TIP1
5. Point (4) was Not Found in Journal File for Feature SPH_TIP1
6. Point (5) was Not Found in Journal File for Feature SPH_TIP1
7. Point (6) was Not Found in Journal File for Feature SPH_TIP1
8. Point (7) was Not Found in Journal File for Feature SPH_TIP1
9. Point (8) was Not Found in Journal File for Feature SPH_TIP1
10. Point (9) was Not Found in Journal File for Feature SPH_TIP1
11. Point (10) was Not Found in Journal File for Feature SPH_TIP1
12. Point (11) was Not Found in Journal File for Feature SPH_TIP1
13. Point (12) was Not Found in Journal File for Feature SPH_TIP1
14. Point (13) was Not Found in Journal File for Feature SPH_TIP1
15. Point (14) was Not Found in Journal File for Feature SPH_TIP1
16. Point (15) was Not Found in Journal File for Feature SPH_TIP1

Any help would be appreciated.

Attached Files
  • I am having the same issue. I had a program that ran without any problems at all. We decided to add a couple of features (1 new plane and 4 new vector points). I initially added the new features at the beginning of the program. We ran the program and received the same error as above (different feature names). The second time we ran it, I moved the features to the very end of the program. Same results. I put in a customer down request to Hexagon and haven't heard a peep from anyone. I'll give them a pass this time because of this pandemic but I am quickly losing my patience. Is there anyone out there that has figured this issue out? I have looked on the knowledge base and FAQs and cannot find anything. Please help!!!
  • It has been a long time since I've worked with this. I do remember that the Journal file is synced with the PRG. If anything is changed, this will happen.

    What is the error in the image? I cannot read it.

    Journal files are date stamped. Do you have an older one the works fine to compare?
  • I've been working with a guy at Hexagon on this issue and apparently it has something to do with 32 bit versus 64 bit and how they react with PC-DMIS NC. Once the feature labels inside the program reach a certain number, they will no longer read back into the PC-DMIS Gateway when they return from the CNC machine. It's frustrating because we had to switch back to running PC-DMIS 2015 and the old NC Server software until they get the 2020 version working properly. Still waiting on that day and it doesn't appear to be coming anytime soon.
  • I can't see that picture that was posted, but looking at the content of the journal file, I am pretty sure that ndirish has it right on the money.

    This looks like it is a known issue with the internal feature IDs in PC-DMIS not being handled correctly.

    The issue is caused because PC-DMIS assigns an internal ID number to every object in the program. The posted CNC program contains these numbers and writes them to the journal file, tagged to each measurment point. This is so the point data winds up back in the right feature when the journal is processed by PC-DMIS.

    This is what goes wrong: PC-DMIS is a 64bit software nowadays, but the variables on the NC control (I am guessing this is a Fanuc) only store 16bit values, so this is where it gets a little complicated and goes wrong. The 64 bit ids get broken up along the path into 4 16bit numers, then reassembled by the time the get back to PC-DMIS


    So the workaround is to keep the ID numbers lower than 65536

    So it means a bit of careful programming.. Cutting and pasting pieces of programs ups these id numbers. Simply deleting old code, from a program will not reset the ID numbers. You have to be carefull to always start with a frech clean program.

    SO OK, I know, this workaround is horrible.

    The good news is that there is a fix in a brand new NC Server that is in beta right now. So there is light at the end of the tunnel.