Looking for new portable CMM

Hi all,
I'm looking for a portable CMM and found some Romers but not sure which one is best fit my need. I need a touch probe CMM, (PC-DMIS software preferred) to measure the mold/die feature mounted on the machine. Accuracy about 0.0002", arm is able to reach 40". I already submit a request to Romer but still want to ask your opinion on any pros and cons of the portable CMM? What machine type you have or suggest, etc...
Currently we have to disassemble the mold/die to put on the CMM to meaaure so I'd liek to avoid that.
Thank you in advance.
Nguyen
Parents
  • I can only say from my own experience I've tested it out on a part that was roughly 25" long in the features I checked. I don't want to start the old Gage R&R on a CMM discussion, but I reran the same part 15 times without moving it and my total variation in any given measurement was .043mm. This was on the arm I spoke of that is not (IMHO) properly utilized. Obviously I avoid the springy part of the steel table, but my probe calibrated at .040mm, and I was using a 6mm probe (which as I stated above I don't like for the flex). I had to use the 6mm for the part I had, and I had to do it quickly because it was an unsanctioned test that I needed to perform so I knew what I was getting off that setup. I had known values on the part I checked from my Cordax RS-150, and those values were within the range of the .043mm (not necessarily the average of the range). Lastly, I tried to measure every hole with the probe coming in at the same angle each time and the same spots in the holes. That's the best I can tell you.
    I am going back up to that facility in a couple days. I'm bringing a tooling ball with me that I've measured on my Cordax. I want to see if I can improve my results using that.

    The accuracy of that 5000 series can definately be better, and I gotta think that with a better arm and a better calibration method I could bring the range of measurements to within a +/-.0006" window. Please noone berate me on my methods here, I was in a hurry and trying to work under the radar. I don't have a lot of support in my strive for excellence at this place.


    Ok. Thanks for the follow up Kami.

    My apologies to Eric.

    So a very careful inspector can hold approximately .0015" would be a more accurate assessment than my +/-.010". I will amend my opinion to reflect that in the future, but I will also note that the OP was inquiring about holding .0002".


Reply
  • I can only say from my own experience I've tested it out on a part that was roughly 25" long in the features I checked. I don't want to start the old Gage R&R on a CMM discussion, but I reran the same part 15 times without moving it and my total variation in any given measurement was .043mm. This was on the arm I spoke of that is not (IMHO) properly utilized. Obviously I avoid the springy part of the steel table, but my probe calibrated at .040mm, and I was using a 6mm probe (which as I stated above I don't like for the flex). I had to use the 6mm for the part I had, and I had to do it quickly because it was an unsanctioned test that I needed to perform so I knew what I was getting off that setup. I had known values on the part I checked from my Cordax RS-150, and those values were within the range of the .043mm (not necessarily the average of the range). Lastly, I tried to measure every hole with the probe coming in at the same angle each time and the same spots in the holes. That's the best I can tell you.
    I am going back up to that facility in a couple days. I'm bringing a tooling ball with me that I've measured on my Cordax. I want to see if I can improve my results using that.

    The accuracy of that 5000 series can definately be better, and I gotta think that with a better arm and a better calibration method I could bring the range of measurements to within a +/-.0006" window. Please noone berate me on my methods here, I was in a hurry and trying to work under the radar. I don't have a lot of support in my strive for excellence at this place.


    Ok. Thanks for the follow up Kami.

    My apologies to Eric.

    So a very careful inspector can hold approximately .0015" would be a more accurate assessment than my +/-.010". I will amend my opinion to reflect that in the future, but I will also note that the OP was inquiring about holding .0002".