Length check off?

Had problems this weekend. Not sure if someone was playing or what? But when I do the (5) Lengths in the uliities for the Romer Arm there is a problem. The 5-6 reading is right (See Below) . Arm locations are the one given in the training manual.

Any help out there?

Thanks

Rob

01/24/11 07:44:58

Serial No: 5130-213

1-2 1016.362
3-4 1016.369
5-6 1015.855
7-8 1016.816
9-10 1016.608

Minimum Length: 1015.855
Maximum Length: 1016.816
Average Length: 1016.402
Range / 2: 0.480
Std. Deviation: 0.335


01/24/11 07:46:57

Serial No: 5130-213

1-2 1016.386
3-4 1016.363
5-6 1015.866
7-8 1016.847
9-10 1016.602

Minimum Length: 1015.866
Maximum Length: 1016.847
Average Length: 1016.413
Range / 2: 0.490
Std. Deviation: 0.338
Parents
  • Do you have results from previous checkouts? When was the last time it was calibrated?

    Sounds like the arm may have been dropped or taken a hit. Look for any obvious signs of damage to the arm and probe.

    Contact romer tech support- you may need to send it in for repair and/or calibration.
Reply
  • Do you have results from previous checkouts? When was the last time it was calibrated?

    Sounds like the arm may have been dropped or taken a hit. Look for any obvious signs of damage to the arm and probe.

    Contact romer tech support- you may need to send it in for repair and/or calibration.