One thing I can't seem to find is if when I reinstall with the Romer Portlock, will PC-DMIS auto recognize the Romer or will I have to go into config and tell it where to look? If I have to tell it where to look, all I have ever seen is COM ports, this Romer is USB and the comp isn't Bluetooth ready so that is off the table... Thanks!
Loading Armspecs in WinRDS 4.1 or higher.
In this new version you can not just copy the folders over manually you must follow this procedure.
1. Power the Arm off and insert the Armspecs Floppy disk.
2. Open the CimCore Arm Utilities short cut on your desktop.
3. Cancel the error message.
4. Select the Config button then the Armspecs Tab.
5. Point the 3 1/2 floppy drive or what ever location you have the Armdata and Armdata.s6x folders located.
* DO NOT SELECT THE ARMDATA FOLDER ONLY POINT TO THE LOCATION IT IS STORED IN.
6. Select OK once then select OK a second time for the Armdata.S6X folder.
Once this is complete you can power the Arm on to connect.
techsupport@romer.com
I tried to select ARMDATA.S6X the second time and it woulnd't let me. PC-DMIS only seems to have probe data for two probes and both are ball probes and I need the scanning probes installed too... Everything worked up until the last selection... I think it may be because there are no .PRB files on the flashdrive... Could that be it?
.PRB files are used only for pcdmis. You have to attach the scanning probe to the arm, if it is detected by winrds it should be listed in pcdmis too, I recommend you to calibrate (qualify) the 15mm ball probe and check your lenght bar to ensure that the arm is in good shape before the rest of probes calibration (follow the instructions in the arm´s documentation).