Bizarre readings in PC-DMIS

Bizarre behavior.  Absolute arm, calibrates fine, length checks, stylus diameter/STDev all fine within RDS software, regardless of length standard or arm orientation (10-20µm at worst).

In PC-DMIS, same arm, same stylus (or master stylus, for that matter), same length standard, length standard reads several millimeters off (3d distance), and is wildly inconsistent.  The readings from PC-DMIS probe readouts window were also noted as each hit was taken, and manual calculation confirms bizarre length readings.  It's as if PC-PMIS is not interfacing with arm's scales correctly.  Once again, RDS software reads fine.

Is there anything we can do short of software re-installation?

Parents Reply
  • When performing the measurement inside PC-DMIS, you need to turn off probe comp, centre the tip in the cone on the scale bar and record a single point (at ball centre), repeat for the next cone position and then do a 3D distance between the two ball centre points.  It should not require an alignment but you do need to make sure the scale bar is held down sufficiently enough so that it doesn't move.

    These are the results of a very quick test I just did myself...