I have 2010MR3 and 2011 on the PC. I can calibrate the CMM-V with no error on 2010MR3. I built the same probe on 2011, and the calibration stop with this kind of message : "Focus detection error in one position".
I have only two choices : retry or cancel. If I retry, it's the same, if I cancel, the probe is not calibrated !!!!!
What can I do ?
(I prefer 2011 for some fixed bugs on tool changer...)
have you made sure all the settings are correct in the settings editor for the cmm-v. When installing newer versions of pc-dmis I have had those setting change on me. I always have to double check the CMM-V/FOCUS PULSE = 32 and the MSI / FRAMEGRABBER TYPE = 301. Not sure what your settings should be. Also make sure that the settings are right in the machine interface setup.
have you made sure all the settings are correct in the settings editor for the cmm-v. When installing newer versions of pc-dmis I have had those setting change on me. I always have to double check the CMM-V/FOCUS PULSE = 32 and the MSI / FRAMEGRABBER TYPE = 301. Not sure what your settings should be. Also make sure that the settings are right in the machine interface setup.