2011 : focus detection error

I have 2010MR3 and 2011 on the PC. I can calibrate the CMM-V with no error on 2010MR3. I built the same probe on 2011, and the calibration stop with this kind of message : "Focus detection error in one position".
I have only two choices : retry or cancel. If I retry, it's the same, if I cancel, the probe is not calibrated !!!!!
What can I do ?
(I prefer 2011 for some fixed bugs on tool changer...)Confused
Parents
  • have you made sure all the settings are correct in the settings editor for the cmm-v. When installing newer versions of pc-dmis I have had those setting change on me. I always have to double check the CMM-V/FOCUS PULSE = 32 and the MSI / FRAMEGRABBER TYPE = 301. Not sure what your settings should be. Also make sure that the settings are right in the machine interface setup.
Reply
  • have you made sure all the settings are correct in the settings editor for the cmm-v. When installing newer versions of pc-dmis I have had those setting change on me. I always have to double check the CMM-V/FOCUS PULSE = 32 and the MSI / FRAMEGRABBER TYPE = 301. Not sure what your settings should be. Also make sure that the settings are right in the machine interface setup.