2011 : focus detection error

I have 2010MR3 and 2011 on the PC. I can calibrate the CMM-V with no error on 2010MR3. I built the same probe on 2011, and the calibration stop with this kind of message : "Focus detection error in one position".
I have only two choices : retry or cancel. If I retry, it's the same, if I cancel, the probe is not calibrated !!!!!
What can I do ?
(I prefer 2011 for some fixed bugs on tool changer...)Confused
Parents
  • The problem has been fixed by our local Hexagon tech, there was a parameter in the constant file which was = 0 (limit of rotary table). Changing it, even if rotary isn't active, solved the problem !!!!!

    Another problem (compatibility between matrox card and win7) has been solved by chaning matrox by IDS card...
    Thanks to Hexagon !
Reply
  • The problem has been fixed by our local Hexagon tech, there was a parameter in the constant file which was = 0 (limit of rotary table). Changing it, even if rotary isn't active, solved the problem !!!!!

    Another problem (compatibility between matrox card and win7) has been solved by chaning matrox by IDS card...
    Thanks to Hexagon !