Good morning!
Running PC-DMIS 2013 MR1 SP5 on Windows 7 x86
----------
I'm embarking on the typical quest of autocalibration routines.
This quest involves an Optiv Performance 443 Camera probe, as well as a plethora of touch probes. I'm moreso concerned with the Camera and 1 probe.
I was able to create a standard autocalibration routine for the single probe using Parameter sets.
However, being an Optiv, I have this big camera I would like to calibrate.
This is a little more complicated unforunately!
However, I am confident. Perhaps overly so?
I am having similar problems as the individual in the aforementioned post. When I calibrate the touch probe, I am using a parameter set with the defaults and MAN+DCC. I am able to start the program and run it, automatically calibrate/qualify the touch probe. However, when I calibrate the camera, the crosshair Hit Target "EM1" to 3 does not show up. I am curious as to why?
I am unable to autocalibrate the camera because of this. To compensate I manually calibrate the camera and have an autocal program for the probes. But I'd like a 1 button solution to both if possible!
Anyone have success with autocalibrating a camera?
Can you try to ASSIGN/V1=QUALTOOLDATA("xyz","the name of the ring"), just to check if PC-DMIS has the location in memory ?
Are you sure to have write rights in the register ?
Can you try to ASSIGN/V1=QUALTOOLDATA("xyz","the name of the ring"), just to check if PC-DMIS has the location in memory ?
Are you sure to have write rights in the register ?
Pretty Sure!
I like to consider myself the resident IT guy (mostly because our external IT services are, suffice to say, lacking) I demanded it.
I can open regedit and add and remove entries at whim.
When I select my tip as the camera, and assign a variable as you have suggested, I am given XYZ coordinates. I am able to insert a move command and get the camera in the center of the ring.
The CMM knows there's a ring there, and knows it's supposed to take points. But when I select MAN+DCC or YES the probe has moved, when using an AUTOCALIBRATE function, I am not able to drag the crosshair to the edge of the ring as normal. THis is my problem.
I would DCC everything, but then that one time it errors out someones going to throw a fit