i have a problem, i have the probe to A0 B0 , then i do the alignment and measure the dimension and everything is fine, but then i need to turn the probe to A0 B90 or 180 and measure, the dimension are not ok, they are totally wrong +-1,5mm more or less, i have all the angles calibrated, what am i doing wrong? Please help
Sorry for mu bad english
Do you use a "L" shape probe ?
You should load your master probe, the calibrate it saying yes, the sphere moved (with rest to theo selected), then load the L probe, and calibrate both angles saying No, the sphere don't moved.
You could create a parameter set and then autocalibrate the probes with it.
Do you use a "L" shape probe ?
You should load your master probe, the calibrate it saying yes, the sphere moved (with rest to theo selected), then load the L probe, and calibrate both angles saying No, the sphere don't moved.
You could create a parameter set and then autocalibrate the probes with it.