DataPage issue

I use datapage V3.33 on a few CMM programs, 3 days ago it was working fine on a particular program and datapage was recording correctly (400 recordings). When running the program I have trace-fields to track work order # and serial number (also saved in datapage). Now when the program runs, datapage records over the previous trace fields (w/o, serial # and stat count #) along with overiding the measured value. I have tried a few things without success, anybody have an idea what can be done or what happend?
Parents
  • -19 sounds like trouble. Ascii dump as much data as you can & create a new database, ascii load the data.
    You say you have a few CMMS. How are you putting data into the database? Is there just 1 database across a network? Are all CMMs using the same database? Unless all paths to the database are exactly the same, there could be corruption issues. If one updates using "S:\database\path" and they are all using the same database, all updates must use "S:\database\path". This includes someone on the local computer looking at the database.
Reply
  • -19 sounds like trouble. Ascii dump as much data as you can & create a new database, ascii load the data.
    You say you have a few CMMS. How are you putting data into the database? Is there just 1 database across a network? Are all CMMs using the same database? Unless all paths to the database are exactly the same, there could be corruption issues. If one updates using "S:\database\path" and they are all using the same database, all updates must use "S:\database\path". This includes someone on the local computer looking at the database.