Hello Everyone,
Occasionally our Leitz (running Quindos 6) 'stalls' with error S_2 CMWAIT CMWT$2 (A COM error I think). A simple 'Gold Enter' and it continues. I would like to trap this error using DFNCDH. My plan is to simply continue from the current point. All of the help in the LANEXT and PALLET manuals seems to want you to go to the start of the element or skip the points completely. Any ideas? Is it as simple as using SKIPPTS with NBR=0?

Thanks

Rob
Parents
  • Please be careful, programming a error handler for an error of unknown cause is a bit risky. The better option would be to determine the cause of the error and fix it, especially if it is a machine communication issue. To progam the CDH to just 'continue' regardless of where the probe may be, could lead to a collision. One must think of the scenario in which this 'stalling' may occur prior to a probe change or with the probe under a part, to just blindly continue is risky.

    Typically, if an error occurs during measurement, the safest would be to always prompt/warn the operator to ensure the probe is free to complete the next move, thus the CDH procedure would involve some sort of operator intervention and acknowledgement, similar to what you have to do now to continue.

    To blindly continue with a CDH is more suited for calculation or analysis commands. For example, if while radius correcting APT's with the RCOR2D command and it errors, a CDH is very well suited to remove erroneous points and continue, without affecting machine movement. CDH's are also commonly used to 'trigger' after the probe has located, or not located a part, during a programmed 'search' probing. But again, to just blindly assume that the probe/machine is always free to continue after a generic error (or random error such as a communication erorr) is risky. Again, it is better to find the cause of this 'communication' error and fix it.

    Please contact us for further help

    Steven Wojciechowski
    Hexagon Metrology

    steven.wojciechowski@hexagonemetrology.com
    401-886-2715

    tech support 800-343-7933
Reply
  • Please be careful, programming a error handler for an error of unknown cause is a bit risky. The better option would be to determine the cause of the error and fix it, especially if it is a machine communication issue. To progam the CDH to just 'continue' regardless of where the probe may be, could lead to a collision. One must think of the scenario in which this 'stalling' may occur prior to a probe change or with the probe under a part, to just blindly continue is risky.

    Typically, if an error occurs during measurement, the safest would be to always prompt/warn the operator to ensure the probe is free to complete the next move, thus the CDH procedure would involve some sort of operator intervention and acknowledgement, similar to what you have to do now to continue.

    To blindly continue with a CDH is more suited for calculation or analysis commands. For example, if while radius correcting APT's with the RCOR2D command and it errors, a CDH is very well suited to remove erroneous points and continue, without affecting machine movement. CDH's are also commonly used to 'trigger' after the probe has located, or not located a part, during a programmed 'search' probing. But again, to just blindly assume that the probe/machine is always free to continue after a generic error (or random error such as a communication erorr) is risky. Again, it is better to find the cause of this 'communication' error and fix it.

    Please contact us for further help

    Steven Wojciechowski
    Hexagon Metrology

    steven.wojciechowski@hexagonemetrology.com
    401-886-2715

    tech support 800-343-7933