Scan Offset Force LSP-S2

All, I currently run PC-DMIS and have the ability to control my scan offset force. I was told by a Quindos user with the same head that they cannot change that value. I do believe you can, if so, how do I tell them to change it?
Thanks!!
Parents
  • you did not elaborate as to what controller you have on your Leitz PMMC. if you have a B4 or B5 controller, you can only change probe parameters (upper force, lower force, trigger force, scan force) by activating one of the parameter sets of the INIFILE in the controller. PC-DMIS may appear to give you the 'control' but the controller will ignore the value(s) sent by the host software (PC-DMIS or Quindos). In Quindos7 you can use the command UseContrlrParSet to activate one of the INIFILE parameter sets.

    contact us at Hexagon if you desire further help
    Steven Wojciechowski 401-886-2715
Reply
  • you did not elaborate as to what controller you have on your Leitz PMMC. if you have a B4 or B5 controller, you can only change probe parameters (upper force, lower force, trigger force, scan force) by activating one of the parameter sets of the INIFILE in the controller. PC-DMIS may appear to give you the 'control' but the controller will ignore the value(s) sent by the host software (PC-DMIS or Quindos). In Quindos7 you can use the command UseContrlrParSet to activate one of the INIFILE parameter sets.