qual. tool locations not updating (v3.7 MR3)

I'm using v3.7 mr3. I qualify on a 1" sphere, "yes" to has it moved. This updates that spheres XYZ definition.
I switch probe configurations, select the same sphere, "no" it hasn't moved. The spheres' definition is not the same as after the first qualify, so the probe heads for an old location. This is a new trick, hasn't given me any trouble before. I tried adding a new "tool". This fixes the problem, but only once. For now,
I have to add a new tool (deleting some old ones) each time I start. What up?
Parents
  • If you have more than one calibration tool stored pcdmis will default to the last one used by that probe.
    You have to change the tool to be used in the probe file.
    You cannot edit a defined calibration tool or you loose all the results relative to that tool.
    The calibration tool data is common across all probes that use that tool.
Reply
  • If you have more than one calibration tool stored pcdmis will default to the last one used by that probe.
    You have to change the tool to be used in the probe file.
    You cannot edit a defined calibration tool or you loose all the results relative to that tool.
    The calibration tool data is common across all probes that use that tool.