HELP error calculation deflection matrix

Error calculation deflection matrix after completing calibration for some tips. what is it?
  • I had this problem intermittently for about a year with a LSPX1M with 100mm extension and 1.5 X 20 probe. DEA controller. Service guy had to tweek a couple of parameters in controller that had to do with timing of wrist rotation or something like that.


    What problem exactly you had? When you say on the controller, do you mean the dim switches?
  • What problem exactly you had? When you say on the controller, do you mean the dim switches?


    The controller is the grey box on casters with cables going to the CMM and your computer. Just about all controller parameters are only accessible via TestSoft. This is a terminal interface that only Hexagon has the keys to.

    The Leitz probe bodies ship in a storage case (currently black plastic), and inside that case there is included a small bottle of Leitz-branded oil. This is what you want to use on the kinematic mounts (the little ball bearings) on the collets. I don't know what the SAE equivalent of that oil is, but you can buy more bottles from Hexagon if you can't find it.

    This is what your OptionProbe statement should look like for a Leitz LSP-X1s:

    OPTIONPROBE/MAXFORCE=0.072,LOWFORCE=0.0096,UPPERFORCE=0.048,TRIGGERFORCE=0.03,
                  PROBEACCURACY=0.1,POSACCURACY=0.1,#RETURNDATA=7,
                  RETURNSPEED=0.4,SCANPNTDENSITY=1,
                  SCANACCELERATION=10,SCANOFFSETFORCE=0.048,
                  PROBINGMODE=DFL,MANFINEPROBING=NO


    I have seen service engineers with plenty of incorrect information...
  • For the LSPX1-S, in setting editor (or regedit) :
    ProbeTriggerForce 0.048 USER_ANALOG_PROBING
    ProbeLowerForce 0.0096 USER_ANALOG_PROBING
    ProbeUpperForce 0.06 USER_ANALOG_PROBING
    ProbeMaxForce 0.12 USER_ANALOG_PROBING
    ScanAcceleration 10 USER_ANALOG_PROBING
    ScanOffsetForce 0.048 USER_ANALOG_PROBING
    ProbeReturnSpeed 0.4 USER_ANALOG_PROBING
    SP6MTXLowerForce_LSPX1S 0.0096 USER_ANALOG_PROBING
    SP6MTXMaxForce_LSPX1S 0.24 USER_ANALOG_PROBING
    SP6MTXUpperForce_LSPX1S 0.12 USER_ANALOG_PROBING
    UseTraxWithSP600 1 Option
    SupportsPRBRDV TRUE LEITZ or FDC
    NewtonToMMConversion_LSPX1S 1.2 LEITZ or FDC
    ConstantSpatialDensity TRUE LEITZ or FDC
    DisableTraxCal FALSE LEITZ or FDC
    VariableHighSpeedScanning TRUE LEITZ or FDC
    SendHSSVectors TRUE LEITZ or FDC
    DisableSP6LowMatrixUpdate FALSE LEITZ or FDC
    DefinedScanOffset 0.03 LEITZ or FDC
    UseAnalogSamplingLatitudeStart 0 ProbeCal
    UseAnalogSamplingNumHits 5 ProbeCal
    DefMode 3 ProbeCal
    UseAnalogSamplingNumLevels 2 ProbeCal
  • This is from the most current Hexagon document I have:


    Attached Files
  • Thank you guys for all your help, I will review all this after lunch, I am almost positive that as far as the probe parameters its all good there.

    I did find the proper oil bottle for the spheres, and I do place OptProbe before the LoadProbe