Does anyone know if the movespeed can cause a TR_FG014, Wrist Error. We have been running the programs at MOVESPEED/15. One tech here is certain that is helping cause the error, so he'll move it back to 10. I still get the error regardless of the movespeed? HELP!!!
Sounds like you have a CW43 or a PHS1 continuous wrist. If this is the case you can put a wrist speed command in PC-DMIS independent of move speed.
Which version of PC-DMIS are you using?
Sounds like you have a CW43 or a PHS1 continuous wrist. If this is the case you can put a wrist speed command in PC-DMIS independent of move speed.
Which version of PC-DMIS are you using?