I have read a number of posts regarding probe calibration and using a Master probe to relate multiple probes. I have a few questions that I did not find the answers in the posts.
Question #1 - If you are using the Master probe for measurement does it affect the relationship of the other probes if you re-calibrate A0/B0 of the Master Probe after answering "N0" to the calibration sphere move. For example - Start of shift the Master probe is calibrated at A0 / B0 and you answer "Yes" the tooling sphere has moved. Now you open a program that uses the master probe for measurement. You want to perform a calibration on the angles that are used in the program. Perform a Marked Used which includes A0/B0. You answer "No" the tool has not moved and perform calibration. Is the relationship of the other probes and angles lost???
Question #2 - Is it necessary to calibrate Master Probe every time the machine is homed??
Meas offsets = theo only when you answer "Yes, the sphere moved" and select reset to theo.
In other cases, the offsets are calculating from the last time when you answered "Yes"...
Meas offsets = theo only when you answer "Yes, the sphere moved" and select reset to theo.
In other cases, the offsets are calculating from the last time when you answered "Yes"...