Auto calibrate question

I am building a program for our operators that specifically calibrates our probes. We use 3 different probes, using a probe changer rack. I am implementing the use of password protection due to the frequent "fat-finger" occurences. I need a way to let operators calibrate as easy as possible. The company i work for requires the operators to calibrate at the beginning of every shift and we run the machines 24/7 with 4 shifts of operators. I have the calibration program built. I just can't find an answer for the meaning of this:

AUTOCALIBRATE/PROBE, PARAMETER_SET=ALL-TIPS-WITH-DEFAULTS, QUALTOOL_MOVED=YES_DCC,
SHOW_SUMMARY=NO, OVERWRITE_RESULTSFILE=YES

the last statement is "OVERWRITE_RESULTSFILE=YES. I assume I want to say YES. However, even when I had NO selected i could see the results files updated when i ran the program. Does anyone know what choosing yes or no does in this situation?
Parents
  • You can print the results as a pdf just like a report. I have the CMM password protected as well. My cal program uses the current date and time as report file name. The probes calibrate automatically with my parameter set. I even use variables to to check for std dev and if its not within my limit, the probe recalibrates. I also included a op comment stating the operator to clean the probes and cal sphere before starting. I don't think I have the actual program offline. Let me try to find it.

    The beauty of this is that the operators can recalibrate the probes as needed and there is a record of doing so. Our op calibrate at the beginning of each shift as well and our cal sphere never moves.

    STARTUP    =ALIGNMENT/START,RECALL:USE_PART_SETUP,LIST=YES
                ALIGNMENT/END
                MODE/DCC
                MOVESPEED/ 80
                MANRETRACT/0.7874
                FORMAT/TEXT,OPTIONS, ,HEADINGS,SYMBOLS, ;NOM,TOL,MEAS,DEV,OUTTOL, ,
                LOADPROBE/VC0022 STAR
                TIP/T1A0B0, SHANKIJK=0, 0, 1, ANGLE=0
                ASSIGN/MM="CMM_"
                ASSIGN/SD=SYSTEMDATE("MM'.'dd'.'yy'_'")
                ASSIGN/ST=SYSTEMTIME("HH':'mm")
                COMMENT/REPT,
    
                "CALIBRATION SERIAL NUMBER:" +MM+SD+ST
    
                ASSIGN/SN=MM+SD+ST
                WORKPLANE/ZPLUS
                ASSIGN/REPORT_LOC="\\\PRISM\quality\CMM\\"
                ASSIGN/REPORT_NAME=REPORT_LOC+SN
    CALIBRATE  =LABEL/
                COMMENT/OPER,NO,FULL SCREEN=NO,AUTO-CONTINUE=NO,
                GENTLY REMOVE PROBE FROM CMM
    
                GENTLY CLEAN TIPS WITH A DRY RAG
    
                PLLACE BACK ON CMM
    
                MAKE SURE RED LIGHT IS ON
    
                CLICK OK WHEN DONE
    AUTOCALIBRATE=LABEL/
                AUTOCALIBRATE/PROBE, PARAMETER_SET=Star_Calibrate, QUALTOOL_MOVED=YES_DCC,
                              SHOW_SUMMARY=YES, OVERWRITE_RESULTSFILE=NO
    BEGIN AUTOCALIBRATE RESULTS FOR PROBE VC0022 STAR USING SET Star_Calibrate
    Probe file=VC0022 STAR    Date=10/3/2017    Time=6:02:09 PM
    Dixie Sphere   CENT X  21.0250 Y   9.8345 Z -12.3495 D 0.7500
    T1A0B0         THEO X   0.0000 Y   0.0000 Z   8.2262 D 0.1969
    T1A0B0         MEAS X  -0.0296 Y  -0.0223 Z   8.1695 D 0.1948 StdDev 0.0002
    T2A0B0         THEO X  -0.9252 Y   0.0000 Z   7.1927 D 0.0787
    T2A0B0         MEAS X  -0.9491 Y  -0.0197 Z   7.1290 D 0.0770 StdDev 0.0002
    T3A0B0         THEO X   0.0000 Y  -0.9252 Z   7.1927 D 0.0787
    T3A0B0         MEAS X  -0.0243 Y  -0.9461 Z   7.1275 D 0.0765 StdDev 0.0003
    T4A0B0         THEO X   0.9252 Y   0.0000 Z   7.1927 D 0.0787
    T4A0B0         MEAS X   0.9035 Y  -0.0166 Z   7.1416 D 0.0769 StdDev 0.0002
    T5A0B0         THEO X   0.0000 Y   0.9252 Z   7.1927 D 0.0787
    T5A0B0         MEAS X  -0.0207 Y   0.9022 Z   7.1400 D 0.0769 StdDev 0.0002
    END AUTOCALIBRATE RESULTS FOR PROBE VC0022 STAR USING SET Star_Calibrate
                LOADPROBE/VC0022 STAR
                ASSIGN/STD1=PROBEDATA("STANDARDDEVIATION","T1A0B0")
                IF_GOTO/STD1>=0.0005,GOTO = AUTOCALIBRATE
                ASSIGN/STD2=PROBEDATA("STANDARDDEVIATION","T2A0B0")
                IF_GOTO/STD2>=0.0005,GOTO = AUTOCALIBRATE
                ASSIGN/STD3=PROBEDATA("STANDARDDEVIATION","T3A0B0")
                IF_GOTO/STD3>=0.0005,GOTO = AUTOCALIBRATE
                ASSIGN/STD4=PROBEDATA("STANDARDDEVIATION","T4A0B0")
                IF_GOTO/STD4>=0.0005,GOTO = AUTOCALIBRATE
                ASSIGN/STD5=PROBEDATA("STANDARDDEVIATION","T5A0B0")
                IF_GOTO/STD5>=0.0005,GOTO = AUTOCALIBRATE
                MOVE/POINT,NORMAL,<12.6,14.8,-6.2541>
                PRINT/REPORT,EXEC MODE=END,$
                  TO_FILE=ON,OVERWRITE=REPORT_NAME,AUTO OPEN REPORT=ON,$
                  TO_PRINTER=OFF,COPIES=1,$
                  TO_DMIS_REPORT=OFF,FILE_OPTION=INDEX,FILENAME=,$
                  REPORT_THEORETICALS=NONE,REPORT_FEATURE_WITH_DIMENSIONS=NO,$
                  TO_EXCEL_OUTPUT=OFF,
                  PREVIOUS_RUNS=KEEP_INSTANCES
    
    
Reply
  • You can print the results as a pdf just like a report. I have the CMM password protected as well. My cal program uses the current date and time as report file name. The probes calibrate automatically with my parameter set. I even use variables to to check for std dev and if its not within my limit, the probe recalibrates. I also included a op comment stating the operator to clean the probes and cal sphere before starting. I don't think I have the actual program offline. Let me try to find it.

    The beauty of this is that the operators can recalibrate the probes as needed and there is a record of doing so. Our op calibrate at the beginning of each shift as well and our cal sphere never moves.

    STARTUP    =ALIGNMENT/START,RECALL:USE_PART_SETUP,LIST=YES
                ALIGNMENT/END
                MODE/DCC
                MOVESPEED/ 80
                MANRETRACT/0.7874
                FORMAT/TEXT,OPTIONS, ,HEADINGS,SYMBOLS, ;NOM,TOL,MEAS,DEV,OUTTOL, ,
                LOADPROBE/VC0022 STAR
                TIP/T1A0B0, SHANKIJK=0, 0, 1, ANGLE=0
                ASSIGN/MM="CMM_"
                ASSIGN/SD=SYSTEMDATE("MM'.'dd'.'yy'_'")
                ASSIGN/ST=SYSTEMTIME("HH':'mm")
                COMMENT/REPT,
    
                "CALIBRATION SERIAL NUMBER:" +MM+SD+ST
    
                ASSIGN/SN=MM+SD+ST
                WORKPLANE/ZPLUS
                ASSIGN/REPORT_LOC="\\\PRISM\quality\CMM\\"
                ASSIGN/REPORT_NAME=REPORT_LOC+SN
    CALIBRATE  =LABEL/
                COMMENT/OPER,NO,FULL SCREEN=NO,AUTO-CONTINUE=NO,
                GENTLY REMOVE PROBE FROM CMM
    
                GENTLY CLEAN TIPS WITH A DRY RAG
    
                PLLACE BACK ON CMM
    
                MAKE SURE RED LIGHT IS ON
    
                CLICK OK WHEN DONE
    AUTOCALIBRATE=LABEL/
                AUTOCALIBRATE/PROBE, PARAMETER_SET=Star_Calibrate, QUALTOOL_MOVED=YES_DCC,
                              SHOW_SUMMARY=YES, OVERWRITE_RESULTSFILE=NO
    BEGIN AUTOCALIBRATE RESULTS FOR PROBE VC0022 STAR USING SET Star_Calibrate
    Probe file=VC0022 STAR    Date=10/3/2017    Time=6:02:09 PM
    Dixie Sphere   CENT X  21.0250 Y   9.8345 Z -12.3495 D 0.7500
    T1A0B0         THEO X   0.0000 Y   0.0000 Z   8.2262 D 0.1969
    T1A0B0         MEAS X  -0.0296 Y  -0.0223 Z   8.1695 D 0.1948 StdDev 0.0002
    T2A0B0         THEO X  -0.9252 Y   0.0000 Z   7.1927 D 0.0787
    T2A0B0         MEAS X  -0.9491 Y  -0.0197 Z   7.1290 D 0.0770 StdDev 0.0002
    T3A0B0         THEO X   0.0000 Y  -0.9252 Z   7.1927 D 0.0787
    T3A0B0         MEAS X  -0.0243 Y  -0.9461 Z   7.1275 D 0.0765 StdDev 0.0003
    T4A0B0         THEO X   0.9252 Y   0.0000 Z   7.1927 D 0.0787
    T4A0B0         MEAS X   0.9035 Y  -0.0166 Z   7.1416 D 0.0769 StdDev 0.0002
    T5A0B0         THEO X   0.0000 Y   0.9252 Z   7.1927 D 0.0787
    T5A0B0         MEAS X  -0.0207 Y   0.9022 Z   7.1400 D 0.0769 StdDev 0.0002
    END AUTOCALIBRATE RESULTS FOR PROBE VC0022 STAR USING SET Star_Calibrate
                LOADPROBE/VC0022 STAR
                ASSIGN/STD1=PROBEDATA("STANDARDDEVIATION","T1A0B0")
                IF_GOTO/STD1>=0.0005,GOTO = AUTOCALIBRATE
                ASSIGN/STD2=PROBEDATA("STANDARDDEVIATION","T2A0B0")
                IF_GOTO/STD2>=0.0005,GOTO = AUTOCALIBRATE
                ASSIGN/STD3=PROBEDATA("STANDARDDEVIATION","T3A0B0")
                IF_GOTO/STD3>=0.0005,GOTO = AUTOCALIBRATE
                ASSIGN/STD4=PROBEDATA("STANDARDDEVIATION","T4A0B0")
                IF_GOTO/STD4>=0.0005,GOTO = AUTOCALIBRATE
                ASSIGN/STD5=PROBEDATA("STANDARDDEVIATION","T5A0B0")
                IF_GOTO/STD5>=0.0005,GOTO = AUTOCALIBRATE
                MOVE/POINT,NORMAL,<12.6,14.8,-6.2541>
                PRINT/REPORT,EXEC MODE=END,$
                  TO_FILE=ON,OVERWRITE=REPORT_NAME,AUTO OPEN REPORT=ON,$
                  TO_PRINTER=OFF,COPIES=1,$
                  TO_DMIS_REPORT=OFF,FILE_OPTION=INDEX,FILENAME=,$
                  REPORT_THEORETICALS=NONE,REPORT_FEATURE_WITH_DIMENSIONS=NO,$
                  TO_EXCEL_OUTPUT=OFF,
                  PREVIOUS_RUNS=KEEP_INSTANCES
    
    
Children
No Data