I am building a program for our operators that specifically calibrates our probes. We use 3 different probes, using a probe changer rack. I am implementing the use of password protection due to the frequent "fat-finger" occurences. I need a way to let operators calibrate as easy as possible. The company i work for requires the operators to calibrate at the beginning of every shift and we run the machines 24/7 with 4 shifts of operators. I have the calibration program built. I just can't find an answer for the meaning of this:
the last statement is "OVERWRITE_RESULTSFILE=YES. I assume I want to say YES. However, even when I had NO selected i could see the results files updated when i ran the program. Does anyone know what choosing yes or no does in this situation?
"and our cal sphere never moves." AUTOCALIBRATE=LABEL/ AUTOCALIBRATE/PROBE, PARAMETER_SET=Star_Calibrate,
QUALTOOL_MOVED=YES_DCC, SHOW_SUMMARY=YES, OVERWRITE_RESULTSFILE=NO
acgarcia just dumb question from beginner, as you don't move cal sphere, red marked code should be
QUALTOOL_MOVED=NO. or ?
Correct. Sphere does not move and the CMM only has a star probe with 5 tips. I have label on the sphere base saying DO NOT MOVE. I have YES_DCC to establish tip 1 of my star probe as the master tip every time they calibrate the probe.
Correct. Sphere does not move and the CMM only has a star probe with 5 tips. I have label on the sphere base saying DO NOT MOVE. I have YES_DCC to establish tip 1 of my star probe as the master tip every time they calibrate the probe.