Auto Calibration

For "Auto Probe Calibration" . so i inserted below command. but while execution it calibrate angles, which are all used in this probe. i need to run only angles which are used in this program. how to set "Parameter Set' to calibrate angles used in this program?
i am using PC-DMIS 2019R2.

STARTUP =ALIGNMENT/START,RECALL:USE_PART_SETUP,LIST=YES
ALIGNMENT/END
MODE/MANUAL

PREHIT/0.1
RETRACT/0.1
MOVESPEED/ 100
SCANSPEED/100
FORMAT/TEXT,OPTIONS, ,HEADINGS,SYMBOLS, ;NOM,TOL,MEAS,DEV,OUTTOL, ,
LOADPROBE/PROBE_4
TIP/T1A0B0, SHANKIJK=0, 0, 1, ANGLE=0
AUTOCALIBRATE/PROBE, PARAMETER_SET=ALL-TIPS-WITH-DEFAULTS, QUALTOOL_MOVED=NO,
CHECK COLLISION=NO, SHOW_SUMMARY=NO, OVERWRITE_RESULTSFILE=NO






  • I have no idea , earliest version I've used is 2015. In my experience which is different for some other users,when ever I created and saved a parameter set, I had to measure it was well so the parameter set would stick. So open the probe dialog box, select your probes, click measure, make your parameter set, save your parameter set, you'll see a prompt saying the set was saved, then click measure and actually calibrate the probes. Someone on this forum told be I didn't' have to do that but again, my experience is that I had to so then I could select it for my autocalibrate program.
  • So I also print a report to our network for record keeping. policy is that the operator must calibrate before each shift. once the calibration is done, the probe moves to the back which is an indicator that the calibration had good results. Otherwise they'd walk up to the CMM and see the operator comment saying the cal was unsuccessful and to seek help.