I guess it's time for me to wake a zombie thread... (but hey, I searched the forum!) (sorry, no actual advanced search options exist here)
We are in the process of coming back online with our machine, which uses a SP-25...
Managment has asked some questions about our calibration process, since we have a "new and different" machine now. Rather than move forward with the default "that's how it's always been done" answer I shall humble myself here on these great forums and hopefully get something more in depth to report.
Seems that the response above is what I was looking for, but just to be clear... (assuming after round 1 new probe calibration has been performed)
To use the system with the best accuracy for scanning features, you must run the default calibration of 5 TTP hits and 8 or so scans per tip angle.
To use the system with the best accuracy for TTP measurements, running a "patrial calibration" that skips the scans and uses 13-25 TTP hits per tip angle will be best.
And these two forms of "calibration" only ever add to each other, not cancel each other out by choosing the wrong option?
99.9999999995% of our programs are TTP only so they understandably care most about that function of the machine.