I'm trying to learn pcdmis 2012. couple weeks ago we hard reset the software. Before that, when an angle finishes the calibration it was avoiding from sphere (at +Z Axis) and continue calibration.
Now it doesnt avoid and probe almost hits the probe changer rack so i hit emergency button and caibrate every angle one by one.
How can i set the avoidance or clearence (i dont know how to call it) at Z axis so i can calibrate multiple angles?
denizborekci , go to OPERATION>CALIBRATE/EDIT>ACTIVE PROBE and then click the SETUP tab. The value you need to change is the clearance distance along the tool shank vector for none continuous wrists. It defaults to 3.5" / 88.9mm
denizborekci , go to OPERATION>CALIBRATE/EDIT>ACTIVE PROBE and then click the SETUP tab. The value you need to change is the clearance distance along the tool shank vector for none continuous wrists. It defaults to 3.5" / 88.9mm