I have a problem with the "SP600" probe head.
The diameters are, regardless of whether they are scanned or with an auto-element circle, it is always exactly 0.010mm smaller than it should.
Tested on a purchased ring with a certificate
Calibration says 1my standard deviation.
Maintenance says everything is ok.
Temperature compensation is off, but the room temperature is 20.5 ° C
Is it possible that the calibration is set incorrectly?
or is it a hardware error?
Once parameters are checked, you have to perform a lower matrix calibration.
Then, you can create a circle scan around the sphere, and look at the difference of radius (it should be close to zero).
You can also check the roundness of the ring (your measured circle), to see if it could be an alignment problem.
Once parameters are checked, you have to perform a lower matrix calibration.
Then, you can create a circle scan around the sphere, and look at the difference of radius (it should be close to zero).
You can also check the roundness of the ring (your measured circle), to see if it could be an alignment problem.