I have a problem with the "SP600" probe head.
The diameters are, regardless of whether they are scanned or with an auto-element circle, it is always exactly 0.010mm smaller than it should.
Tested on a purchased ring with a certificate
Calibration says 1my standard deviation.
Maintenance says everything is ok.
Temperature compensation is off, but the room temperature is 20.5 ° C
Is it possible that the calibration is set incorrectly?
or is it a hardware error?
clean the tip and the sphere
select the L50D8 (length 50 mm - diameter 8 mm : it's the "standard tip" for the SP600)
select "reset to theo", and "yes, the sphere moved"
calibrate by selecting "lower matrix", then re calibrate.
I'm not sure that it will change lot the diameter, maybe more the circularity.
For the sphere, you should look to the stem, the diameter is often write on it.