Upper Management is looking for answers/Justification on probe breakage

Hello ALL GURU's:
Here's a great post for a lot of input on WWGD (What would Gurus Do)?
Lets start by saying there are NOT an excessive amount of probes being broken yearly. I would estimate less that $500.00 - $700.00 yearly max, if that.
However, The uppers want justification.
Now, it has been suggested to slow the rapid in the machine settings, which would cause more cycle time on check parts that are in a Pallet setting.
There is a CRASH attachment from Renishaw. Anyone currently use these?. And do they work. Say on a .7x20mm probe or a 1mmx27mm probe?.
In our shop, we have 4 Cmm's that all have touch probes and 2 have vision as well. What would a normal (realistic) probe consumption be expected
at your place?. Mind you that the specific probes that we have in question are 1.5x30mm. 1mmx27.5 and .7x20mm. Sometimes it may just be human error (can't help that).
Maybe a slight misload, part has been machined or molded not into spec. Other times, maybe doing a manual align and you forget the rapid is on and "Snap". It happens.
I was asked to figure the FORCE in which a probe breaks. Well, honestly what other factors go into that one?. Are you rapiding, your prehit/retract too close and you touch the part and snap?.
Seeing as we are a high volume molding/machine shop with multiple parts at any given time to inspect via DCC. What is the best course of action to go ZERO probe breakage a year?.
This would not include Set ups, Prove Outs as well.
I am just looking for some insight on this issue from a lot of people who know way more than me!. Hence, picking the brains of the Guru's. ANY input would be greatly appreciated and if you want to keep it short and sweet. great. I don't want to consume too much of your time writing.
Thank You All Slight smile

Parents
  • I put this together, feel free to steal/use as you see fit.

    NOTE-- Rename *.PDF to *.XLSX before opening. --We are restricted from uploading xlsx files apparently.

    Probe breakage is inevitable, no matter how much you implement to isolate it.
    You are better offf tracking the frequency, and producing a spares inventory. I have gotten to the point that i can safely say we burn through about 9 1x27.5's on average each year. Boom in the budget for next year, and we have a full set of production spares at the ready for each year.
    I'm also tracking bigger ticket items like probe modules, bodies and heads, as well as maintenance activities to log all unscheduled maintenance and costs including HMA / "MyCare" current bullhonkey.
    I manually insert the mycare/HMA costs in the analysis tab.
Reply
  • I put this together, feel free to steal/use as you see fit.

    NOTE-- Rename *.PDF to *.XLSX before opening. --We are restricted from uploading xlsx files apparently.

    Probe breakage is inevitable, no matter how much you implement to isolate it.
    You are better offf tracking the frequency, and producing a spares inventory. I have gotten to the point that i can safely say we burn through about 9 1x27.5's on average each year. Boom in the budget for next year, and we have a full set of production spares at the ready for each year.
    I'm also tracking bigger ticket items like probe modules, bodies and heads, as well as maintenance activities to log all unscheduled maintenance and costs including HMA / "MyCare" current bullhonkey.
    I manually insert the mycare/HMA costs in the analysis tab.